Calculation of the measured value
The measured value is calculated in the following steps between reading the ADC and outputting the measured value:
Offset correction
Formula: Calculation of the measured value 1:
Otherwise: Calculation of the measured value 2:
CoE Index
Meaning
Unit
Raw Value
XADC
0x800E:01
0x9000:01*)
Output value of the A/D converter
Range: 0..262,144 (18 bits) + sign in the respective active measuring range
Unit:
[Digits / VFSV] or
[Digits / AFSV]
Example:
Instantaneous value = 218 / FSV [V or A] * electrical value
BZ
0x800E:02
0x9000:02*)
Current ADC Zero value
BZCal
0x800F:20
0x9000:03*)
ADC Zero at the time of calibration
YZ
Measured value after ADC offset correction
Note BZ: Contains the measured value of the last ADC conversion with Zero Compensation or (if Zero Compensation Interval = 0) the value from 0x800F:1F.
Filtering
Formula: Calculation of the measured value 3:
Otherwise: Calculation of the measured value 4:
CoE Index
Meaning
Unit
Enable Filter
0x8000:04
YF
Measured value after averaging
50/60 Hz filtering
Formula at 60 Hz: Calculation of the measured value 5:
Formula at 50 Hz: Calculation of the measured value 6:
CoE Index
Meaning
Unit
Frequency
0x8000:0A
50/60Hz interference suppression
BF
0x800F:1D
60 Hz offset
Unit:
[Digits / VFSV] or
[Digits / AFSV]
Example:
AF
0x800F:1E
60 Hz gain
Y60
Measured value after adaptation of the 60 Hz integration time
Vendor calibration
The previously determined measured value "raw value" is now offset by the vendor calibration values by default so that the channel meets the specified value.
Formula: Calculation of the measured value 7:
Otherwise: Calculation of the measured value 8:
CoE Index
Meaning
Unit
Enable Vendor calibration
0x8000:01
Enables SAI Vendor data 0x800F; enabled by default
BK
0x800F
(odd subindices)
Vendor Data: Offset
Unit:
[Digits / VFSV] or
[Digits / AFSV]
Example:
AK
0x800F
(even subindices)
Vendor Data: Gain
YH
Measured value after vendor calibration
User calibration
If necessary, the measured value can be changed on the application side.
The terminal has one user calibration value pair for each measuring range and measuring mode, i.e. 14 pairs in total. To increase the resolution of the gain value, the "+1" (0x4000) is omitted.
As gain values smaller than 1 can also occur, the gain value is provided with an offset of 1/8. For a gain of 1 the gain value must be 0x2000. For gain G = 1 the following applies:
Calculation of the measured value 9:
For other values, "G" must be replaced by the desired gain factor.
Before calibration, the presentation of the values should be set to "right-aligned": CoE index 0x8000:1D (Presentation = 1).
Formula: Calculation of the measured value 10:
Otherwise: Calculation of the measured value 11:
CoE Index
Meaning
Unit
Enable User calibration
0x8000:02
Enables SAI vendor data 0x8002; default disabled
BA
0x8002
(odd subindices)
User Data: Offset
Unit:
[Digits / VFSV] or
[Digits / AFSV]
Example:
AA
0x8002
(even subindices)
User Data: Gain
YA
Measured value after user calibration
AC system offset
This internal calculation is mandatory in the data stream and cannot be changed or switched off. This is used for linearization with small AC values.
Formela: Calculation of the measured value 12:  for Calculation of the measured value 13:
If condition is not met: Calculation of the measured value 14:
CoE Index
Meaning
Unit
XSO
0x800F:1F
AC system offset
Y
Measured value after AC system offset
Unit:
[Digits / VFSV] or
[Digits / AFSV]
Example:
Scaling
During scaling, the previously calculated measured value is converted to the desired output format. The scaling contains the powers of ten selected by the range.
All gain factors must therefore correspond to 1/8 to obtain a total factor of 1. The value 8192 (0x2000) results in a total factor of 1
Formula: Calculation of the measured value 15:
If condition is not met: Calculation of the measured value 16:
CoE Index
Meaning
Unit
Presentation
0x8000:1D
FScal
YScale
0x6000:11
Measured value after scaling
Presentation e.g. in scaled:
1 bit per 1 µV / 1 µA
*) from Firmware 06 / Revision -0021