ELM3002-02x5/ ELM3002-03x5/ ELM3002-04x5 - Technical data
Technical data |
ELM3002-02x5/ -03x5/ -04x5 |
Analog inputs |
2 channels |
Time relation between channels |
Simultaneous conversion of all channels in the terminal, synchronous conversion between terminals if Distributed Clocks is used |
ADC conversion method |
ΔΣ (deltaSigma) with internal 8 Msamples/s sample rate |
Ground reference |
Differential |
Input filter limit frequency for hardware |
Before AD converter: Hardware low-pass -3 dB @ 330 kHz, type: Butterworth 4th order In the AD converter after conversion: adjustable low-pass filter The ramp-up time/ settling time/ delay caused by the filtering will be considered within the Distributed Clocks-Timestamp |
Resolution |
24-bit (including sign) |
Connection technology |
2-wire |
Sampling rate |
20 µs/50 kSps per channel, fixed Then free down sampling by Firmware via decimation factor |
Oversampling |
1...100 can be selected |
Supported EtherCAT cycle time (dependent on operation mode) |
Distributed Clocks: min. 100 µs, max. 10 ms FrameTriggered/Synchronous: min. 200 µs, max. 100 ms FreeRun: not yet supported |
Connection diagnosis |
Internal analog ground AGND |
Not accessible |
Surge voltage protection of the inputs, based on AGND (internal ground) |
Internal power supply |
via E-bus |
Current consumption via E-bus |
typ. 460 mA |
Current consumption of power contacts |
Thermal power dissipation |
typ. 3 W |
Dielectric strength – destruction limit |
See information on electrical isolation |
Recommended operating voltage range for compliance with specification |
Common data | ELM3002-02x5/ -03x5/ -04x5 |
Distributed Clocks | Yes, with oversampling n = 1…100, accuracy << 1 μs |
Special features | ExtendedRange 107 % and 400%, free numeric filter, TrueRMS, integrator/differentiator, non-linear scaling, PeakHold |
Functional diagnosis 1) | Yes |
Electrical isolation | Reinforced insulation in accordance with EN 61010-2-030, EN60664-1 6 kV DC, 5 sec ramp, 2 sec hold production test Type test: 13 kV DC, 1 min Max. operating voltage range (continuous, operation without CAT measurement category, i.e. without transient overvoltages): |
Measurement category / overvoltage category | 1000 V CAT II; according to EN 61010-2-030 in preparation: 1000 V CAT III, 600 V CAT IV; according to EN 61010-2-030 Notes: |
Configuration | via the EtherCAT Master, e.g. TwinCAT |
Note to cable length | - |
1) see chapters Self-test and self-test report and Connection test/switchable connection diagnosis
2) see notes to potential groups within chapter Power supply, potential groups
3) Explanation UDiff, USE, UCM
Basic mechanical | ELM3002-02x5/ -03x5/ -04x5 |
Connection type | 4 mm safety laboratory socket, touch-protected, maintenance plug |
Dimensions (W x H x D) | 30 mm x 100 mm x 102 mm, see chapter Housing, ELM3xxx documentation |
Mounting | on 35 mm rail conforms to EN 60715 |
Note mounting | Plugs not within scope of delivery, see chapter Notes on connection technology, ELM3xxx documentation |
Weight | Approx. 350 g |
Permissible ambient temperature range during operation | 0…+55 °C (-25…+60 °C approval in preparation) |
Permissible ambient temperature range during storage | -40…+85 °C |
Environmental data | ELM3002-02x5/ -03x5/ -04x5 |
Permissible operating altitude range | 0 up to 2000 m according to EN 61010-1) |
Relative humidity | max. 95%, no condensation |
Protection class | IP 20 |
Installation position | Variable (Caution with the installation position type “Mounting rail at top, connection side faces downwards”, please see ELM3002-0xx5 – Safety and operating instructions) |
Pollution level | 2 or better according to EN 61010-1) |
Normative data | ELM3002-02x5/ -03x5/ -04x5 |
Vibration-/shock resistance | Conforms to EN 60068-2-6 / EN 60068-2-27 |
EMC-immunity/ emission | Conforms to EN 61000-6-2 / EN 61000-6-4 |
Approvals/ markings *) | CE |
EMC notes | ESD contact discharges conforming to EN61000-6-4 onto the terminal enclosure can lead to measurement deviations up to ± 1%FSV on all channels. |