Measurement ±360 V
ELM3002-02x5 (channel 1, channel 2), ELM3002-04x5 (channel 1 only)
Measurement mode
±360 V
Measuring range, nominal
-360 …+360 V
Measuring range, end value (FSV)
360 V
Measuring range, technically usable
-386.547…+386.547 V
PDO resolution (including sign)
24 bit
16 bit 2)
PDO LSB (Extended Range)
46.08 µV
11.79.. mV
PDO LSB (Legacy Range)
42.91.. µV
10.98.. mV
Measurement mode
±360 V
Basic accuracy: Measuring deviation at 23°C, with averaging 6)
< ±0.015 %FSV, < ±150 ppmFSV typ.
< ±0.05 V typ.
Extended basic accuracy: Measuring deviation at 0…55°C, with averaging 6)
< ±0.0305, < ±305 ppmFSV typ.
< ±109.8 mV typ.
Offset/Zero point deviation (at 23°C)
EOffset
< 130 ppmFSV
Gain/scale/amplification deviation (at 23°C)
EGain
< 65 ppm
Non-linearity over the whole measuring range
ELin
< 20 ppmFSV
Repeatability
ERep
< 8 ppmFSV
Temperature coefficient
TcGain
< 8 ppm/K typ.
TcOffset
< 2 ppmFSV/K typ.
< 0.72 mV/K typ.
Largest short-term deviation during a specified electrical interference test
< ±0.1 %FSV typ.
Input impedance ±Input
(internal resistance)
Differential typ. approx. > 10 MΩ || < 1nF
CommonMode typ. against SGND: tbd.
2) The analog measurement is always done with 24 bits, in 16-bit mode the eight least significant bits are cut off. For further information see chapter “Commissioning”/ “Process data overview”
6) Calculated value according to the equation for estimating usability at the given ambient temperature point or estimating usability over the specified ambient temperature range in operation (Tambient) as stated in chapter "General information on measuring accuracy/measurement uncertainty". In real use, for example at a relatively constant ambient temperature Tambient, a lower (better) achievable uncertainty is attained. A specific calculation according to chapter "General information on measuring accuracy/measurement uncertainty" is recommended, especially if the instrument allows a wider ambient temperature range in operation than 0...55 °C.
ELM3002-02x5 (channel 1, channel 2), ELM3002-04x5 (channel 1 only) (50 ksps)
Noise (without filtering)
ENoise, PtP
< 60 ppmFSV
< 469 digits
< 21.6 mV
ENoise, RMS
< 5 ppmFSV
< 39 digits
< 1.8 mV
Max. SNR
> 106 dB
Noisedensity@1kHz
< 11.38 Measurement ±360 V 1:
Noise (with 50/60 Hz FIR filter 5))
ENoise, PtP
< 8 ppmFSV
< 63 digits
< 2.88 mV
ENoise, RMS
< 1.5 ppmFSV
< 12 digits
< 0.54 mV
Max. SNR
> 116.5 dB
Crosstalk (without filter), typ.
DC: > tbd. dB
50/60 Hz: > tbd. dB
1 kHz: > tbd. dB
Crosstalk (with 50/60 Hz FIR filter 5)), typ.
DC: > tbd. dB
50/60 Hz: > tbd. dB
1 kHz: > tbd. dB
5) the 50/60 Hz filter setting must be selected via the settings in the CoE depending on the local mains frequency (50 or 60 Hz); the respective nominal mains frequency then applies as the measuring point for common mode rejection or crosstalk.
