ELM334x - Technical data
Technical data | ELM3344-00xx | ELM3348-00xx |
Analog inputs | 4 channel (differential) | 8 channel (differential) |
Time relation between channels to each other | Simultaneous conversion of all channels in the terminal (multiplex), synchronous conversion between terminals, if Distributed Clocks will be used. Timestamp each channel, typ. sampling offset related to channel 1: | |
Ch.1: 0 µs | Ch.1: 0 µs | |
ADC conversion method | ΔΣ (Delta-Sigma) with internal sample rate 8 Msps | |
Cutoff frequency input filtering hardware ELM Features/ Firmware filtering concept) | Before AD converter: Within ADC after conversion: Low pass: -3 dB @ 10.9 kHz type sinc3/average filter The ramp-up time/ settling time/ delay caused by the filtering will be considered within the Distributed Clocks-Timestamp. | |
Resolution | 24 bit (including sign) | |
Connection technology | 2-wire | |
Sampling rate (per channel) | 1 ms/ 1 ksps | |
Free down sampling by Firmware via decimation factor | ||
Oversampling | 1…25 selectable | |
Supported EtherCAT cycle time (depending on the operation mode) | Distributed Clocks: FrameTriggered/Synchron: FreeRun: not yet supported | |
Connection diagnosis | Wire break | |
Internal power supply | Via E-bus | |
Current consumption E-bus | typ. 220 mA | typ. 260 mA |
Current consumption power contacts | - | |
Thermal power dissipation | typ. 2 W | |
Dielectric strength - destruction limit | max. permitted short-term/permanent voltage between the contact points: ±30 V | |
Recommended operation voltage range to compliance with specification | max. permitted voltage at +Input, −Input, related to the internal analog ground (AGND) (“operation as intended“): ±12.5 V | |
Common data | ELM3344-00xx | ELM3348-00xx |
Distributed Clocks | Yes, with Oversampling n = 1…25, accuracy << 1 μs | |
Special features | AGND (of the ELM334x-00x0) leaded through, common mode measurement, freely configurable numeric filters, TrueRMS, integrator/differentiator, non-linear scaling, PeakHold | |
Functional diagnosis 1) | Yes | |
Electrical isolation channel/channel 2) | no | |
Electrical isolation channel/E-Bus 2) | Functional insulation, 707 V DC (type test) | |
Electrical isolation channel/SGND 2) | Functional insulation, 707 V DC (type test) | |
Configuration | via the EtherCAT Master, e.g. TwinCAT | |
Note to cable length | Signal cable lengths to the sensor / encoder over 3 m must be shielded, the shield design must be in line with the state of the art and be effective. For larger cable lengths > 30 m, a suitable surge protection should be provided if appropriate interference could affect the signal cable. | |
1) see chapters Self-test and self-test report and Connection test/switchable connection diagnosis
2) see notes to potential groups within chapter Power supply, potential groups
Basic mechanical | ELM3344-00xx | ELM3348-00xx |
Connection type | ELM3344-00x0 / ELM3348-00x0: | |
ELM3344-00x3 / ELM3348-00x3: | ||
Dimensions (W x H x D) | See chapter Housing | |
Mounting | on 35 mm rail conforms to EN 60715 | |
Note mounting | ELM334x-00x0: plug within scope of delivery, | |
ELM334x-00x3: plug not within scope of delivery, | ||
Weight | approx. 350 g | |
Permissible ambient temperature range during operation | 0…+55 °C | |
Permissible ambient temperature range during storage | -25…+85 °C | |
Environmental data | ELM3344-00xx | ELM3348-00xx |
Permissible operating altitude range | 0 up to 2000 m (derating at higher altitudes on request) | |
Relative humidity | max. 95%, no condensation | |
Protection class | IP 20 | |
Installation position | Variable | |
Normative data | ELM3344-00xx | ELM3348-00xx |
Vibration-/shock resistance | Conforms to EN 60068-2-6 / EN 60068-2-27 | |
EMC-resistance / emission | Conforms to EN 61000-6-2 / EN 61000-6-4 | |
Approvals/ markings *) | CE, UKCA, cULus_55, EAC | |
EMC notes | In case of push-in and mini-TC connectors, ESD air discharges conforming to EN 61000-6-4 into the connectors or to the lines connected there can lead to measurement deviations up to ±FSV within the respective channel or to other channels by crosstalk. Peak voltages (surge) conforming to EN 61000-6-2 into the Up supply (power contact) at set connection “Connect Up- to GNDA“ or “Connect Up- to AGND“ within CoE (F800:01) can lead to measurement deviations up to ±FSV. | |
