Technical data
All values are typical values over the entire temperature range, unless stated otherwise.
Supply voltages | |
Connection | See EtherCAT P connection |
US nominal voltage |
24 VDC (-15 % / +20 %) |
US sum current: IS,sum |
max. 3 A |
Current consumption from US |
100 mA |
Rated voltage UP | 24 VDC (-15 % / +20 %) |
UP sum current: IP,sum |
max. 3 A |
Current consumption from UP | None. UP is only forwarded. |
Accelerometers | |
Sensor type | Two 3-axis sensors / offset by 90° |
Resolution 1) 2) | Measured values: 4 mg (default) Raw values: 10-bit (default) |
Representation 1) | Measured values: 1 mg / LSB Raw values: 10-bit in 16-bit (left aligned) |
Measuring range 1) | ±2 g / ±4 g / ±8 g / ±16 g selectively |
Sampling rate | 200 Hz to 5 kHz |
1) Unit of measurement: 1 g = 9.81 m/s2 (acceleration of gravity). 1 mg = 1/1000 g.
2) The resolution depends on the parameterization of the box. See section Resolution.
Housing data | |
Dimensions W x H x D | 30 mm x 126 mm x 26.5 mm (without connectors) |
Weight | approx. 165 g |
Material |
PA6 (polyamide) |
Environmental conditions | |
Ambient temperature during operation | -25 … +60 °C |
Ambient temperature during storage | -40 … +85 °C |
Vibration resistance, shock resistance | conforms to EN 60068-2-6 / EN 60068-2-27 |
Protection class | IP65, IP66, IP67 (conforms to EN 60529) |
Approvals / markings | |
Approvals / markings *) | CE, cULus |
Resolution
The resolution of measured values and raw values depends on the parameters "Measuring range" and "Sampling rate". The table below shows how these parameters influence the resolution:
Measuring range | Sampling rate / | Resolution | |
Raw values | Measured values | ||
±2 g | ≤ 1 kHz / | 10-bit | 4 mg |
±4 g | 8 mg | ||
±8 g | 16 mg | ||
±16 g | 48 mg | ||
±2 g | > 1 kHz / | 8-bit | 16 mg |
±4 g | 32 mg | ||
±8 g | 64 mg | ||
±16 g | 192 mg | ||
The setting of the "Measuring range" and "Sampling rate" parameters is described in the chapter Settings.
Additional tests
The devices have undergone the following additional tests:
Test | Explanation |
Vibration | 10 frequency sweeps in 3 axes |
5 Hz < f < 60 Hz displacement 0.35 mm, constant amplitude | |
60.1 Hz < f < 500 Hz acceleration 5 g, constant amplitude | |
Shocks | 1000 shocks in each direction, in 3 axes |
35 g, 11 ms |
