Technical data
All values are typical values over the entire temperature range, unless stated otherwise.
EtherCAT
Connection
B23 ENP hybrid connector
Electrical isolation
500 V
Supply voltages
Connection
US rated voltage
24 V DC (-15%/+20%)
US sum current: IS,sum
max. 28 A at 45°C 1)
Current consumption from US
150 mA
+ output currents of the digital outputs
+ sensor power supply
UP nominal voltage range
18…48 V DC
UP sum current: IP,sum
max. 28 A at 45 °C 1)
Current consumption from UP
= Output currents for the motors
UP undervoltage detection
18 V
UP overvoltage detection
60 V
1) This value corresponds to the current carrying capacity of the connectors.
Motor channels
Number
2
Motor type
3-phase BLDC roller motors
Connection
2 x M8 socket, 4-pin, A-coded
Cable length
max. 30 m
Motor inductance
min. 200 µH
Rotary encoder
No rotary encoder needed
Motor voltage
18…48 V DC from the supply voltage UP
Continuous current per channel
3.5 Arms
Peak current per channel
12.0 Arms for approx. 1 s
Rotating field velocity
599 Hz 2)
PWM clock frequency
16 kHz
Current controller frequency
32 kHz
2) Do not confuse the rotating field velocity with the motor velocity or the roller velocity.
Digital inputs/outputs
Number
8
Connections
4 x M8 socket
Cable length
max. 30 m
Sensor power supply US1
24 V DC from the control voltage US
max. 0.5 A, short-circuit proof
Input specification
Characteristics
Type 3 according to EN 61131-2, compatible with type 1
Input filter
10 µs
Output specification
Nominal voltage
24 V DC (-15 % / +20 %) from the control voltage US
Output current
max. 0.5 A per output, individually short-circuit proof
Housing data
Weight
750 g
Installation position
variable
Environmental conditions
Ambient temperature during operation
-25…+60 °C
-25…+55 °C according to cURus
Ambient temperature during storage
-40…+85 °C
Degree of pollution
2, for use under UL conditions
Vibration resistance, shock resistance
conforms to EN 60068-2-6 / EN 60068-2-27
Additional checks
EMC immunity / emission
conforms to EN 61000-6-2 / EN 61000-6-4
Protection rating
IP65, IP66, IP67 (according to EN 60529)
Approvals/markings
Approvals/markings *)
CE, cURus