Technical data
All values are typical values over the entire temperature range, unless stated otherwise.
EtherCAT  | |
Connection  | 2 x M8 socket, 4-pin, A-coded, shielded  |
Electrical isolation  | 500 V  |
Distributed Clocks  | yes  |
Minimum cycle time  | 100 µs  |
Supply voltages  | |
Connection  | Input: M8 connector, 4-pin, A-coded Downstream connection: M8 socket, 4-pin, A-coded  |
									 US nominal voltage  |
									 24 VDC (-15 % / +20 %)  |
US sum current: IS,sum  | max. 4 A  |
									 Current consumption from US  |
									 120 mA  |
Rated voltage UP  | 24 VDC (-15 % / +20 %)  |
UP sum current: IP,sum  | max. 4 A  |
Current consumption from UP  | The measuring bridge supply voltage Uref is generated internally from UP.  |
Measuring bridge connection  | |
M12 socket, 5-pin, A-coded  | |
Signal type  | Resistor bridge, strain gauge (SG)  |
Number of analog inputs  | 2, for 1 bridge circuit in full bridge technology.  |
Cable length  | max. 30 m between box and measuring bridge  |
Measured value resolution  | 24-bit, 32-bit representation  |
Conversion time  | 0.1 … 250 ms, configurable  |
Conversion rate  | 4 ... 10,000 samples/s (reciprocal of the conversion time)  |
Measuring error  | < ±0.01 % 1) for the calculated load value based on the load end value with 10 V power supply and 20 mV bridge voltage (therefore nominal characteristic value for SG: 2 mV/V), self-calibration active, 50 Hz filter active.  |
Measuring range UD  | max. -27 … +27 mV 2) Recommended: -25 … +25 mV nominal voltage  |
Measuring range Uref  | max. -13.8 … +13.8 V 2)  |
Supported nominal characteristic values  | any, resolution of the parameter: 0.01 µV/V Recommended: 0.5 … 2.5 mV/V  |
min. strain gauge resistance  | Parallel use of strain gauges only recommended with suitable strain gauges  |
Input filter (hardware) cut-off frequency  | 10 kHz low-pass (-3 dB, see filter notes)  |
Filter (software)  | Preset 50 Hz, configurable:  |
Internal resistance  |
Measuring bridge supply voltage (sensor power supply)  | Uref = 10 V (generated internally from UP) max. 350 mA  |
1) External influences such as Temperature drift and RF interference can increase the measuring error.
2) See chapter Voltage measurement.
Housing data  | |
Dimensions W x H x D  | 60 mm x 126 mm x 26.5 mm (without plug connectors)  |
Weight  | approx. 250 g  |
Installation position  | variable  |
									 Material  |
									 PA6 (polyamide)  |
Environmental conditions  | |
Ambient temperature during operation  | -25 … +60 °C  |
Ambient temperature during storage  | -40 … +85 °C  |
Vibration resistance, shock resistance  | conforms to EN 60068-2-6 / EN 60068-2-27  |
EMC immunity / emission  | conforms to EN 61000-6-2 / EN 61000-6-4  |
Protection class  | IP65, IP66, IP67 (conforms to EN 60529)  |
Approvals/markings  | |
Approvals/markings *)  | CE, cURus  |
Additional tests
The devices have undergone the following additional tests:
Test  | Explanation  |
Vibration  | 10 frequency sweeps in 3 axes  |
5 Hz < f < 60 Hz displacement 0.35 mm, constant amplitude  | |
60.1 Hz < f < 500 Hz acceleration 5 g, constant amplitude  | |
Shocks  | 1000 shocks in each direction, in 3 axes  |
35 g, 11 ms  |