TC measurement with Beckhoff terminals
Thermocouple specification and conversion
Temperature measurement with thermocouples generally comprises three steps:
All three steps can take place locally in the Beckhoff measuring device. Device-based transformation can be disabled if the conversion is to take place in the higher-level control system. Depending on the device type, several thermocouple conversions are available, which differ in terms of their software implementation.
For Beckhoff thermocouple measuring devices this means that
TC measurement with Beckhoff terminals 1:
Data for the sensor types in the following table
The values for the sensor types listed in the following table are shown here merely for informative purposes as an orientation aid. All data are given without guarantee and must be cross-checked against the data sheet for the respective sensor employed.
The thermocouple measurement consists of a chain of measuring and computing elements that affect the attainable measurement deviation:
TC measurement with Beckhoff terminals 2:Fig.90: Concatenation of the uncertainties in temperature measurement with thermocouples
The given voltage specification is decisive for the achievable temperature measuring accuracy. It is applied to the possible thermocouple types in the following.
On account of
detailed temperature specification tables are not given below, but rather
Notes on the calculation of detailed specifications
If further specifications are of interest, they can or must be calculated from the values given in the voltage specification.
The sequence:
The numerical values used in the following three examples are for illustration purposes. The specification values given in the technical data remain authoritative.
Sample 1:
Basic accuracy of an ELM3704 at 35 °C ambient, measurement of 400 °C with thermocouple type K, without noise and aging influences:
Tmeasuring point = 400 °C
MW = UType K, 400°C = 16.397 mV
TC measurement with Beckhoff terminals 5:
= 100.196 ppmFSV
FVoltage(Umeasuring point) = 100.196 ppmFSV · 80 mV = 8.016 µV
ΔUperK(Tmeasuring point) = (U(401 °C) - U(400 °C)) / (1 °C) = 42.243 µV/°C
FCJC, T = tbd
FCJC, U(Tmeasuring point) = tbd °C · 42.243 µV/°C = tbd µV
FVoltage+CJC = tbd
FELM3704@35°C, type K, 400°C = (Fvoltage+CJC µV) / (42.243 µV/°C) ≈ tbd °C (means ±tbd °C)
Sample 2:
Consideration of the repeatability alone under the above conditions:
MW=Umeasuring point (400 °C) = 16.397 mV
FSingle = 20 ppmFSV
FVoltage = 20 ppmFSV · 80 mV = 1.6 µV
FCJC, single = tbd °C
FCJC, Single, U(Tmeasuring point) = tbd °C · 42.243 µV/°C = tbd µV
FVoltage+CJC = tbd
FTemp(Umeasuring point) = (Fvoltage+CJC µV) / (42.243 µV/°C) ≈ tbd °C (means ±tbd °C)
Sample 3:
Consideration of the RMS noise alone without filter under the above conditions:
FSingle = 37 ppmFSV
FVoltage = 37 ppmFSV · 80 mV = 2.96 µV
FVoltage+CJC = tbd