Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection
Notes
By a user-side adjustment with a connected bridge sensor, the measurement uncertainity related to gain and offset error can be significant reduced.
The integrated switcheable shunt resistor can be used to generate a predictable detuning or, in case of deviation, a correction factor.
To calculate the quarter-bridge:
Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 1:
Fig.103: Connection of the quarter bridge
Explanation:
The strain relationship (µStrain, µε) is as follows:
Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 2:
For the quarter-bridge, N=1 always applies.
The relationship between UBridge/UExc and ∆R1 is non-linear:
Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 3:
Fig.104: Relationship between UBridge/UExc and ∆R1
The ELM350x devices apply internal linearization so that the output is already linearized
Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 4:
since the internal calculation is based on UExc'.
Measurement mode
Measuring bridge/StrainGauge SG 1/4-bridge 120 Ω 2/3-wire
32 mV/V
8 mV/V
4 mV/V (comp.) 5)
2 mV/V (comp.) 5)
Measuring range, nominal
±32 mV/V [corresponds to ±64,000 µε at K=2]
120 ±15.36 Ω
±8 mV/V [corresponds to ±16,000 µε at K=2]
120 ±3.84 Ω
±4 mV/V [corresponds to ±8,000 µε at K=2]
120 ±1.92 Ω
±2 mV/V [corresponds to ±4,000 µε at K=2]
120 ±0.96 Ω
Measuring range, end value (FSV)
32 mV/V
8 mV/V
4 mV/V
2 mV/V
Measuring range, technically usable
±34.359... mV/V
±8.589... mV/V
±4.294... mV/V
±2.147... mV/V
PDO resolution
24 bit (including sign)
PDO LSB (Extended Range)
0.128 ppm
4.096 nV/V
0.128 ppm
1.024 nV/V
0.128 ppm
0.512 nV/V
0.128 ppm
0.256 nV/V
PDO LSB (Legacy Range)
0.119… ppm
3.814.. nV/V
0.119… ppm
0.9535 nV/V
0.119… ppm
0.47675 nV/V
0.119… ppm
0.238375 nV/V
Basic accuracy: Measuring deviation at 23°C, with averaging, typ. 2) 6)
without Offset
< ±0.026 %FSV
< ±260 ppmFSV
< ±8.3 µV/V
< ±0.08 %FSV
< ±800 ppmFSV
< ±6.4 µV/V
< ±0.16 %FSV
< ±1600 ppmFSV
< ±6.4 µV/V
< ±0.32 %FSV
< ±3200 ppmFSV
< ±6.4 µV/V
incl. Offset
< ±0.1 %FSV
< ±1000 ppmFSV
< ±32 µV/V
< ±0.4 %FSV
< ±4000 ppmFSV
< ±32 µV/V
< ±0.8 %FSV
< ±8000 ppmFSV
< ±32 µV/V
< ±1.6 %FSV
< ±16000 ppmFSV
< ±32 µV/V
Extended basic accuracy: Measuring deviation at 0…55°C, with averaging, typ. 2) 6)
without Offset
< ±0.164 %FSV
< ±1640 ppmFSV
< ±52.5 µV/V
< ±0.565 %FSV
< ±5650 ppmFSV
< ±45.2 µV/V
< ±1.1295 %FSV
< ±11295 ppmFSV
< ±45.2 µV/V
< ±2.2585 %FSV
< ±22585 ppmFSV
< ±45.2 µV/V
incl. Offset
< ±0.1995 %FSV
< ±1995 ppmFSV
< ±63.8 µV/V
< ±0.718 %FSV
< ±7180 ppmFSV
< ±57.4 µV/V
< ±1.436 %FSV
< ±14360 ppmFSV
< ±57.4 µV/V
< ±2.872 %FSV
< ±28720 ppmFSV
< ±57.4 µV/V
Offset/Zero point deviation (at 23°C) 4)
EOffset
< 960 ppmFSV
< 3920 ppmFSV
< 7840 ppmFSV
< 15680 ppmFSV
Gain/scale/amplification deviation (at 23°C)
EGain
< 160 ppm
< 520 ppm
< 1040 ppm
< 2080 ppm
Non-linearity over the whole measuring range
ELin
< 200 ppmFSV
< 600 ppmFSV
< 1200 ppmFSV
< 2400 ppmFSV
Repeatability, over 24 h, with averaging
ERep
< 25 ppmFSV
< 100 ppmFSV
< 200 ppmFSV
< 400 ppmFSV
Common-mode rejection ratio (without filter) 3)
tbd.
tbd.
tbd.
tbd.
Common-mode rejection ratio (with 50/60 Hz FIR filter) 1) 3)
tbd.
tbd.
tbd.
tbd.
Temperature coefficient, typ.
TcGain
< 20 ppm/K
< 48 ppm/K
< 96 ppm/K
< 192 ppm/K
TcOffset
< 50 ppmFSV/K
< 1.6 µV/V/K
< 180 ppmFSV/K
< 1.44 µV/V/K
< 360 ppmFSV/K
< 1.44 µV/V/K
< 720 ppmFSV/K
< 1.44 µV/V/K
Largest short-term deviation during a specified electrical interference test
tbd. %FSV
tbd. %FSV
tbd. %FSV
tbd. %FSV
Input impedance ±Input 1
Differential
tbd.
tbd.
tbd.
tbd.
CommonMode
tbd.
tbd.
tbd.
tbd.
Input impedance ±Input 2
3-wire
Differential
tbd.
tbd.
tbd.
tbd.
CommonMode
tbd.
tbd.
tbd.
tbd.
ELM3502 (20 ksps)
Measurement mode
32 mV/V
8 mV/V
4 mV/V (comp.) 5)
2 mV/V (comp.) 5)
Noise (without filtering, at 23°C)
ENoise, PtP
< 310 ppmFSV
< 2422 digits
< 9.92 µV/V
< 1200 ppmFSV
< 9375 digits
< 9.60 µV/V
< 2400 ppmFSV
< 18750 digits
< 9.60 µV/V
< 4800 ppmFSV
< 37500 digits
< 9.60 µV/V
ENoise, RMS
< 50 ppmFSV
< 391 digits
< 1.60 µV/V
< 200 ppmFSV
< 1563 digits
< 1.60 µV/V
< 400 ppmFSV
< 3125 digits
< 1.60 µV/V
< 800 ppmFSV
< 6250 digits
< 1.60 µV/V
Max. SNR
> 86.0 dB
> 74.0 dB
> 68.0 dB
> 61.9 dB
Noisedensity@1kHz
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 5:
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 6:
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 7:
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 8:
Noise (with 50/60 Hz FIR filter, at 23°C) 1)
ENoise, PtP
< 24 ppmFSV
< 188 digits
< 0.77 µV/V
< 72 ppmFSV
< 563 digits
< 0.58 µV/V
< 144 ppmFSV
< 1125 digits
< 0.58 µV/V
< 288 ppmFSV
< 2250 digits
< 0.58 µV/V
ENoise, RMS
< 4.0 ppmFSV
< 31 digits
< 0.13 µV/V
< 12.0 ppmFSV
< 94 digits
< 0.10 µV/V
< 24.0 ppmFSV
< 188 digits
< 0.10 µV/V
< 48.0 ppmFSV
< 375 digits
< 0.10 µV/V
Max. SNR
> 108.0 dB
> 98.4 dB
> 92.4 dB
> 86.4 dB
ELM3504 (10 ksps)
Measurement mode
32 mV/V
8 mV/V
4 mV/V (comp.) 5)
2 mV/V (comp.) 5)
ENoise, PtP
< 285 ppmFSV
< 2227 digits
< 9.12 µV/V
< 1000 ppmFSV
< 7813 digits
< 8.00 µV/V
< 2000 ppmFSV
< 15625 digits
< 8.00 µV/V
< 4000 ppmFSV
< 31250 digits
< 8.00 µV/V
ENoise, RMS
< 50 ppmFSV
< 391 digits
< 1.60 µV/V
< 150 ppmFSV
< 1172 digits
< 1.20 µV/V
< 300 ppmFSV
< 2344 digits
< 1.20 µV/V
< 600 ppmFSV
< 4688 digits
< 1.20 µV/V
Max. SNR
> 86.0 dB
> 76.5 dB
> 70.5 dB
> 64.4 dB
Noisedensity@1kHz
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 9:
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 10:
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 11:
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 12:
ENoise, PtP
< 20 ppmFSV
< 156 digits
< 0.64 µV/V
< 60 ppmFSV
< 469 digits
< 0.48 µV/V
< 120 ppmFSV
< 938 digits
< 0.48 µV/V
< 240 ppmFSV
< 1875 digits
< 0.48 µV/V
ENoise, RMS
< 4.0 ppmFSV
< 31 digits
< 0.13 µV/V
< 12.0 ppmFSV
< 94 digits
< 0.10 µV/V
< 24.0 ppmFSV
< 188 digits
< 0.10 µV/V
< 48.0 ppmFSV
< 375 digits
< 0.10 µV/V
Max. SNR
> 108.0 dB
> 98.4 dB
> 92.4 dB
> 86.4 dB
Measurement mode
Measuring bridge/StrainGauge SG 1/4-bridge 350 Ω 2/3-wire
32 mV/V
8 mV/V
4 mV/V (comp.) 5)
2 mV/V (comp.) 5)
350 ±44.8 Ω
350 ±11.2 Ω
350 ±5.6 Ω
350 ±2.8 Ω
32 mV/V
8 mV/V
4 mV/V
2 mV/V
±34.359... mV/V
±8.589... mV/V
±4.294... mV/V
±2.147... mV/V
PDO resolution
0.128 ppm
4.096 nV/V
0.128 ppm
1.024 nV/V
0.128 ppm
0.512 nV/V
0.128 ppm
0.256 nV/V
0.119… ppm
3.814.. nV/V
0.119… ppm
0.9535 nV/V
0.119… ppm
0.47675 nV/V
0.119… ppm
0.238375 nV/V
without Offset
< ±0.022 %FSV
< ±220 ppmFSV
< ±7 µV/V
< ±0.08 %FSV
< ±800 ppmFSV
< ±6.4 µV/V
< ±0.16 %FSV
< ±1600 ppmFSV
< ±6.4 µV/V
< ±0.32 %FSV
< ±3200 ppmFSV
< ±6.4 µV/V
incl. Offset
< ±0.1 %FSV
< ±1000 ppmFSV
< ±32 µV/V
< ±0.4 %FSV
< ±4000 ppmFSV
< ±32 µV/V
< ±0.8 %FSV
< ±8000 ppmFSV
< ±32 µV/V
< ±1.6 %FSV
< ±16000 ppmFSV
< ±32 µV/V
without Offset
< ±0.0995 %FSV
< ±995 ppmFSV
< ±31.8 µV/V
< ±0.3715 %FSV
< ±3715 ppmFSV
< ±29.7 µV/V
< ±0.7425 %FSV
< ±7425 ppmFSV
< ±29.7 µV/V
< ±1.485 %FSV
< ±14850 ppmFSV
< ±29.7 µV/V
incl. Offset
< ±0.144 %FSV
< ±1440 ppmFSV
< ±46.1 µV/V
< ±0.5565 %FSV
< ±5565 ppmFSV
< ±44.5 µV/V
< ±1.113 %FSV
< ±11130 ppmFSV
< ±44.5 µV/V
< ±2.2255 %FSV
< ±22255 ppmFSV
< ±44.5 µV/V
EOffset
< 970 ppmFSV
< 3920 ppmFSV
< 7840 ppmFSV
< 15680 ppmFSV
EGain
< 120 ppm
< 440 ppm
< 880 ppm
< 1760 ppm
ELin
< 180 ppmFSV
< 650 ppmFSV
< 1300 ppmFSV
< 2600 ppmFSV
ERep
< 25 ppmFSV
< 100 ppmFSV
< 200 ppmFSV
< 400 ppmFSV
tbd.
tbd.
tbd.
tbd.
tbd.
tbd.
tbd.
tbd.
TcGain
< 12 ppm/K
< 50 ppm/K
< 100 ppm/K
< 200 ppm/K
TcOffset
< 30 ppmFSV/K
< 0.96 µV/V/K
< 110 ppmFSV/K
< 0.88 µV/V/K
< 220 ppmFSV/K
< 0.88 µV/V/K
< 440 ppmFSV/K
< 0.88 µV/V/K
tbd. %FSV
tbd. %FSV
tbd. %FSV
tbd. %FSV
Differential
tbd.
tbd.
tbd.
tbd.
CommonMode
tbd.
tbd.
tbd.
tbd.
3-wire
Differential
tbd.
tbd.
tbd.
tbd.
CommonMode
tbd.
tbd.
tbd.
tbd.
ELM3502 (20 ksps)
Measurement mode
32 mV/V
8 mV/V
4 mV/V (comp.) 5)
2 mV/V (comp.) 5)
ENoise, PtP
< 320 ppmFSV
< 2500 digits
< 10.24 µV/V
< 1200 ppmFSV
< 9375 digits
< 9.60 µV/V
< 2400 ppmFSV
< 18750 digits
< 9.60 µV/V
< 4800 ppmFSV
< 37500 digits
< 9.60 µV/V
ENoise, RMS
< 55 ppmFSV
< 430 digits
< 1.76 µV/V
< 200 ppmFSV
< 1563 digits
< 1.60 µV/V
< 400 ppmFSV
< 3125 digits
< 1.60 µV/V
< 800 ppmFSV
< 6250 digits
< 1.60 µV/V
Max. SNR
> 85.2 dB
> 74.0 dB
> 68.0 dB
> 61.9 dB
Noisedensity@1kHz
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 13:
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 14:
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 15:
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 16:
ENoise, PtP
< 18 ppmFSV
< 141 digits
< 0.58 µV/V
< 72 ppmFSV
< 563 digits
< 0.58 µV/V
< 144 ppmFSV
< 1125 digits
< 0.58 µV/V
< 288 ppmFSV
< 2250 digits
< 0.58 µV/V
ENoise, RMS
< 3.0 ppmFSV
< 23 digits
< 0.10 µV/V
< 12.0 ppmFSV
< 94 digits
< 0.10 µV/V
< 24.0 ppmFSV
< 188 digits
< 0.10 µV/V
< 48.0 ppmFSV
< 375 digits
< 0.10 µV/V
Max. SNR
> 110.5 dB
> 98.4 dB
> 92.4 dB
> 86.4 dB
ELM3504 (10 ksps)
Measurement mode
32 mV/V
8 mV/V
4 mV/V (comp.) 5)
2 mV/V (comp.) 5)
ENoise, PtP
< 290 ppmFSV
< 2266 digits
< 9.28 µV/V
< 1000 ppmFSV
< 7813 digits
< 8.00 µV/V
< 2000 ppmFSV
< 15625 digits
< 8.00 µV/V
< 4000 ppmFSV
< 31250 digits
< 8.00 µV/V
ENoise, RMS
< 50 ppmFSV
< 391 digits
< 1.60 µV/V
< 160 ppmFSV
< 1250 digits
< 1.28 µV/V
< 320 ppmFSV
< 2500 digits
< 1.28 µV/V
< 640 ppmFSV
< 5000 digits
< 1.28 µV/V
Max. SNR
> 86.0 dB
> 75.9 dB
> 69.9 dB
> 63.9 dB
Noisedensity@1kHz
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 17:
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 18:
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 19:
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 20:
ENoise, PtP
< 15 ppmFSV
< 117 digits
< 0.48 µV/V
< 50 ppmFSV
< 391 digits
< 0.40 µV/V
< 100 ppmFSV
< 781 digits
< 0.40 µV/V
< 200 ppmFSV
< 1563 digits
< 0.40 µV/V
ENoise, RMS
< 3.0 ppmFSV
< 23 digits
< 0.10 µV/V
< 9.0 ppmFSV
< 70 digits
< 0.07 µV/V
< 18.0 ppmFSV
< 141 digits
< 0.07 µV/V
< 36.0 ppmFSV
< 281 digits
< 0.07 µV/V
Max. SNR
> 110.5 dB
> 100.9 dB
> 94.9 dB
> 88.9 dB
Measurement mode
Measuring bridge/StrainGauge SG 1/4-bridge 1 kΩ 2/3-wire
32 mV/V
8 mV/V
4 mV/V (comp.) 5)
2 mV/V (comp.) 5)
1000 ±128 Ω
1000 ±32 Ω
1000 ±16 Ω
1000 ±8 Ω
32 mV/V
128 Ω
8 mV/V
32 Ω
4 mV/V
16 Ω
2 mV/V
8 Ω
±34.359... mV/V
±8.589... mV/V
±4.294... mV/V
±2.147... mV/V
PDO resolution
0.128 ppm
4.096 nV/V
0.128 ppm
1.024 nV/V
0.128 ppm
0.512 nV/V
0.128 ppm
0.256 nV/V
0.119… ppm
3.814.. nV/V
0.119… ppm
0.9535 nV/V
0.119… ppm
0.47675 nV/V
0.119… ppm
0.238375 nV/V
without Offset
< ±0.02 %FSV
< ±200 ppmFSV
< ±6.4 µV/V
< ±0.065 %FSV
< ±650 ppmFSV
< ±5.2 µV/V
< ±0.13 %FSV
< ±1300 ppmFSV
< ±5.2 µV/V
< ±0.26 %FSV
< ±2600 ppmFSV
< ±5.2 µV/V
incl. Offset
< ±0.1 %FSV
< ±1000 ppmFSV
< ±32 µV/V
< ±0.4 %FSV
< ±4000 ppmFSV
< ±32 µV/V
< ±0.8 %FSV
< ±8000 ppmFSV
< ±32 µV/V
< ±1.6 %FSV
< ±16000 ppmFSV
< ±32 µV/V
without Offset
< ±0.1855 %FSV
< ±1855 ppmFSV
< ±59.4 µV/V
< ±0.6975 %FSV
< ±6975 ppmFSV
< ±55.8 µV/V
< ±1.3945 %FSV
< ±13945 ppmFSV
< ±55.8 µV/V
< ±2.7885 %FSV
< ±27885 ppmFSV
< ±55.8 µV/V
incl. Offset
< ±0.2205 %FSV
< ±2205 ppmFSV
< ±70.6 µV/V
< ±0.8415 %FSV
< ±8415 ppmFSV
< ±67.3 µV/V
< ±1.683 %FSV
< ±16830 ppmFSV
< ±67.3 µV/V
< ±3.366 %FSV
< ±33660 ppmFSV
< ±67.3 µV/V
EOffset
< 980 ppmFSV
< 3940 ppmFSV
< 7880 ppmFSV
< 15760 ppmFSV
EGain
< 105 ppm
< 305 ppm
< 610 ppm
< 1220 ppm
ELin
< 165 ppmFSV
< 560 ppmFSV
< 1120 ppmFSV
< 2240 ppmFSV
ERep
< 25 ppmFSV
< 120 ppmFSV
< 240 ppmFSV
< 480 ppmFSV
tbd.
tbd.
tbd.
tbd.
tbd.
tbd.
tbd.
tbd.
TcGain
< 13 ppm/K
< 25 ppm/K
< 50 ppm/K
< 100 ppm/K
TcOffset
< 60 ppmFSV/K
< 1.92 µV/V/K
< 230 ppmFSV/K
< 1.84 µV/V/K
< 460 ppmFSV/K
< 1.84 µV/V/K
< 920 ppmFSV/K
< 1.84 µV/V/K
tbd. %FSV
tbd. %FSV
tbd. %FSV
tbd. %FSV
Differential
tbd.
tbd.
tbd.
tbd.
CommonMode
tbd.
tbd.
tbd.
tbd.
3-wire
No usage of this input in this mode
Differential
tbd.
tbd.
tbd.
tbd.
CommonMode
tbd.
tbd.
tbd.
tbd.
ELM3502 (20 ksps)
Measurement mode
32 mV/V
8 mV/V
4 mV/V (comp.) 5)
2 mV/V (comp.) 5)
ENoise, PtP
< 400 ppmFSV
< 3125 digits
< 12.80 µV/V
< 1350 ppmFSV
< 10547 digits
< 10.80 µV/V
< 2700 ppmFSV
< 21094 digits
< 10.80 µV/V
< 5400 ppmFSV
< 42188 digits
< 10.80 µV/V
ENoise, RMS
< 65 ppmFSV
< 508 digits
< 2.08 µV/V
< 240 ppmFSV
< 1875 digits
< 1.92 µV/V
< 480 ppmFSV
< 3750 digits
< 1.92 µV/V
< 960 ppmFSV
< 7500 digits
< 1.92 µV/V
Max. SNR
> 83.7 dB
> 72.4 dB
> 66.4 dB
> 60.4 dB
Noisedensity@1kHz
< 0.03 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 21:
< 0.03 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 22:
< 0.03 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 23:
< 0.03 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 24:
ENoise, PtP
< 60 ppmFSV
< 469 digits
< 1.92 µV/V
< 240 ppmFSV
< 1875 digits
< 1.92 µV/V
< 480 ppmFSV
< 3750 digits
< 1.92 µV/V
< 960 ppmFSV
< 7500 digits
< 1.92 µV/V
ENoise, RMS
< 10.0 ppmFSV
< 78 digits
< 0.32 µV/V
< 40.0 ppmFSV
< 313 digits
< 0.32 µV/V
< 80.0 ppmFSV
< 625 digits
< 0.32 µV/V
< 160.0 ppmFSV
< 1250 digits
< 0.32 µV/V
Max. SNR
> 100.0 dB
> 88.0 dB
> 81.9 dB
> 75.9 dB
ELM3504 (10 ksps)
Measurement mode
32 mV/V
8 mV/V
4 mV/V (comp.) 5)
2 mV/V (comp.) 5)
ENoise, PtP
< 350 ppmFSV
< 2734 digits
< 11.20 µV/V
< 820 ppmFSV
< 6406 digits
< 6.56 µV/V
< 1640 ppmFSV
< 12813 digits
< 6.56 µV/V
< 3280 ppmFSV
< 25625 digits
< 6.56 µV/V
ENoise, RMS
< 70 ppmFSV
< 547 digits
< 2.24 µV/V
< 140 ppmFSV
< 1094 digits
< 1.12 µV/V
< 280 ppmFSV
< 2188 digits
< 1.12 µV/V
< 560 ppmFSV
< 4375 digits
< 1.12 µV/V
Max. SNR
> 83.1 dB
> 77.1 dB
> 71.1 dB
> 65.0 dB
Noisedensity@1kHz
< 0.03 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 25:
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 26:
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 27:
< 0.02 Measurement SG 1/4 bridge (quarter-bridge) 2/3-wire connection 28:
ENoise, PtP
< 85 ppmFSV
< 664 digits
< 2.72 µV/V
< 48 ppmFSV
< 375 digits
< 0.38 µV/V
< 96 ppmFSV
< 750 digits
< 0.38 µV/V
< 192 ppmFSV
< 1500 digits
< 0.38 µV/V
ENoise, RMS
< 14.0 ppmFSV
< 109 digits
< 0.45 µV/V
< 8.0 ppmFSV
< 63 digits
< 0.06 µV/V
< 16.0 ppmFSV
< 125 digits
< 0.06 µV/V
< 32.0 ppmFSV
< 250 digits
< 0.06 µV/V
Max. SNR
> 97.1 dB
> 101.9 dB
> 95.9 dB
> 89.9 dB
1) the 50/60 Hz filter setting must be selected via the settings in the CoE depending on the local mains frequency (50 or 60 Hz); the respective nominal mains frequency then applies as the measuring point for common mode rejection or crosstalk.
2) In real bridge measurement, an offset adjustment is usually carried out after installation. The given offset specification of the terminal is therefore practically irrelevant. Therefore, specification values with and without offset are given here. In practice, the offset component can be eliminated by the functions ELM Features and also ELM Features of the terminal or in the controller by a higher-level tare function. The offset deviation of a bridge measurement over time can change, therefore Beckhoff recommends a regular offset adjustment or careful observation of the change.
3) Values related to a common mode interference between SGND and internal ground.
4) The offset specification does not apply to 2-wire operation, since the offset is increased on the device side. Therefore, a system-side offset adjustment is recommended, see Tare- or ZeroOffset function. The final targeting basic acuuracy within the 2-wire operation is mainingly dependent by the quality of this system-side offset adjustment.
5) The channel measures electrically to 8 mV/V, but displays its measured value scaled to 2 or 4 mV/V. The Compensated function facilitates measurement of low levels even with high offset.
6) Calculated value according to the equation for estimating usability at the given ambient temperature point or estimating usability over the specified ambient temperature range in operation (Tambient) as stated in chapter "General information on measuring accuracy/measurement uncertainty". In real use, for example at a relatively constant ambient temperature Tambient, a lower (better) achievable uncertainty is attained. A specific calculation according to chapter "General information on measuring accuracy/measurement uncertainty" is recommended, especially if the instrument allows a wider ambient temperature range in operation than 0...55 °C.
Notice
Transition resistances of the connection contacts
The transition resistance values of the connection contacts affect the measurement. The measuring accuracy can be further increased by a user-side adjustment with the signal connection plugged in.
The temperature sensitivity of the terminal and thus of the measurement setup can be reduced if an external, more temperature-stable supplementary resistor is used for operation of the terminal in half-bridge or even full-bridge mode instead of the internal supplementary resistor for quarter-bridge mode.