Measurement IEPE ±20 mV
Measurement mode | ±20 mV | ||
Measuring range, nominal | -20…+20 mV | ||
Measuring range, end value (FSV) | 20 mV | ||
Measuring range, technically usable | -21.474…+21.474 mV | ||
PDO resolution (including sign) | 24 bit | 16 bit 2) | |
PDO LSB (Extended Range) | 2.56 nV | 655.36 nV | |
PDO LSB (Legacy Range) | 2.384.. nV | 610.37.. nV | |
Input impedance ±Input 1 (internal resistance) | Differential typ. 2 MΩ || 1 nF CommonMode typ. 10 nF against SGND | ||
2) The analog measurement is always done with 24 bits, in 16-bit mode the eight least significant bits are cut off. For further information see chapter “Commissioning”/ “Process data overview”
Basic accuracy: Measuring deviation at 23°C, with averaging 1) 6) | < ±0.04 % = 400 ppmFSV typ. < ±8 µV typ. | |
Extended basic accuracy: Measuring deviation at 0…55°C, with averaging 1) 6) | < ±0.077 %, < ±770 ppmFSV typ. < ±15.4 µV typ. | |
Offset/Zero point deviation (at 23°C) 1) | EOffset | < 380 ppmFSV |
Gain/scale/amplification deviation (at 23°C) 1) | EGain | < 60 ppm |
Non-linearity over the whole measuring range 1) | ELin | < 100 ppmFSV |
Repeatability, over 24 h, with averaging 1) | ERep | < 25 ppmFSV |
Temperature coefficient 1) | TcGain | < 4 ppm/K typ. |
TcOffset | < 20 ppmFSV/K typ. < 0.4 µV/K typ. | |
Largest short-term deviation during a specified electrical interference test | Value to follow | |
1) Valid for ELM3602-00x0 from HW08, ELM360x-00x2 from HW09, ELM3604-00x0 from HW10; specifications of predecessor-HW on request
6) Calculated value according to the equation for estimating usability at the given ambient temperature point or estimating usability over the specified ambient temperature range in operation (Tambient) as stated in chapter "General information on measuring accuracy/measurement uncertainty". In real use, for example at a relatively constant ambient temperature Tambient, a lower (better) achievable uncertainty is attained. A specific calculation according to chapter "General information on measuring accuracy/measurement uncertainty" is recommended, especially if the instrument allows a wider ambient temperature range in operation than 0...55 °C.
ELM3602 (50 ksps)
Measurement mode | ±20 mV | |||
Noise (without filtering) | ENoise, PtP | < 1200 ppmFSV | < 9375 digits | < 24 µV |
ENoise, RMS | < 200 ppmFSV | < 1563 digits | < 4 µV | |
Max. SNR | > 74 dB | |||
Noisedensity@1kHz | < 40 | |||
Noise (with 50/60 Hz FIR filter) 5) | ENoise, PtP | < 140 ppmFSV | < 1094 digits | < 2.8 µV |
ENoise, RMS | < 23 ppmFSV | < 180 digits | < 0.46 µV | |
Max. SNR | > 92.8 dB | |||
Crosstalk (without filter) | DC: tbd. dB typ. | 50/60 Hz: tbd. dB typ. | 1 kHz: tbd. dB typ. | |
Crosstalk (with 50/60 Hz FIR filter) 5) | DC: tbd. dB typ. | 50/60 Hz: tbd. dB typ. | 1 kHz: tbd. dB typ. | |
ELM3604 (20 ksps)
Noise (without filtering) | ENoise, PtP | < 700 ppmFSV | < 5469 digits | < 14 µV |
ENoise, RMS | < 120 ppmFSV | < 938 digits | < 2.4 µV | |
Max. SNR | > 78.4 dB | |||
Noisedensity@1kHz | < 24 | |||
Noise (with 50/60 Hz FIR filter) 5) | ENoise, PtP | < 80 ppmFSV | < 625 digits | < 1.6 µV |
ENoise, RMS | < 16 ppmFSV | < 125 digits | < 0.32 µV | |
Max. SNR | > 95.9 dB | |||
Crosstalk (without filter) | DC: < -100 dB typ. | 50/60 Hz: < -100 dB typ. | 1 kHz: < -75 dB typ. | |
Crosstalk (with 50/60 Hz FIR filter) 5) | DC: < -100 dB typ. | 50/60 Hz: < -120 dB typ. | 1 kHz: < -120 dB typ. | |
5) the 50/60 Hz filter setting must be selected via the settings in the CoE depending on the local mains frequency (50 or 60 Hz); the respective nominal mains frequency then applies as the measuring point for common mode rejection or crosstalk.
Fig.20: Representation ±20 mV measurement range|
Note: In Extended Range Mode the Underrange/Overrange display in the PDO status has the character of an information/warning when the nominal measuring range is exceeded, i.e. no Error is displayed in the PDO status and LED. If the technical measuring range is also exceeded, Error = TRUE is also displayed. The detection limit for Underrange/Overrange Error can be set in the CoE. In Legacy Range mode, an Underrange/Overrange event also leads to an Error in the PDO status. |
Frequency response: See data within ±10V measurement mode
