Measurement IEPE 0..10 V
Measurement mode | 0…10 V | ||
Measuring range, nominal | 0…10 V | ||
Measuring range, end value (FSV) | 10 V | ||
Measuring range, technically usable | 0…+10.737 V | ||
PDO resolution (unsigned) | 23 bit | 15 bit 2) | |
PDO LSB (Extended Range) | 1.28 µV | 327.68 µV | |
PDO LSB (Legacy Range) | 1.192.. µV | 305.18.. µV | |
Input impedance ±Input 1 (internal resistance) | Differential typ. 550 kΩ || 11 nF | ||
2) The analog measurement is always done with 24 bits, in 16-bit mode the eight least significant bits are cut off. For further information see chapter “Commissioning”/ “Process data overview”
3) For IEPE measurement applies: The input voltage must not fall below -5 V with respect to GND, the measuring accuracy is then no longer given. This means a measurement down to -10 V with respect to GND is only possible if at the same time an offset of at least +5 V is applied, as is usual with IEPE supply.
Specific data (not valid for ELM3704-10x1/ -10x2, preliminary data in cursive format)
Measurement mode | 0…10 V | |||
Basic accuracy: Measuring deviation at 23°C, with averaging 1) 6) | < ±0.05 %FSV
Extended basic accuracy: Measuring deviation at 55°C, with averaging 1) 6) | < ±0.113 %FSV
Offset/Zero point deviation (at 23°C) 1) | EOffset | < 300 ppmFSV | ||
Gain/scale/amplification deviation (at 23°C) 1) | EGain | < 100 ppm | ||
Non-linearity over the whole measuring range 1) | ELin | < 380 ppmFSV | ||
Repeatability, over 24 h, with averaging 1) | ERep | < 10 ppmFSV | ||
Noise (without filtering) | ENoise, PtP | < 75 ppmFSV | < 586 digits | < 0.75 mV |
ENoise, RMS | < 13 ppmFSV | < 98 digits | < 0.13 mV | |
Max. SNR | > 98.1 dB | |||
Noisedensity@1kHz | < 1.77 | |||
Noise (with 50/60 Hz FIR filter) 5) | ENoise, PtP | < 18 ppmFSV | < 141 digits | < 0.18 mV |
ENoise, RMS | < 3 ppmFSV | < 23 digits | < 30 µV | |
Max. SNR | > 110.5 dB | |||
Temperature coefficient 1) | TcGain | < 30 ppm/K typ. | ||
TcOffset | < 10 ppmFSV/K typ. | |||
Crosstalk (without filter) | DC: >115 dB typ. | 50/60 Hz: >105 dB typ. | 1 kHz: >80 dB typ. | |
Crosstalk (with 50/60 Hz FIR filter) 5) | DC: >115 dB typ. | 50/60 Hz: >115 dB typ. | 1 kHz: >115 dB typ. | |
Largest short-term deviation during a specified electrical interference test | ±0.03 % = 300 ppmFSV typ. | |||
1) Valid for ELM3704-00x1 from HW00, ELM370x-00x0 from HW01; specifications of predecessor-HW on request
5) the 50/60 Hz filter setting must be selected via the settings in the CoE depending on the local mains frequency (50 or 60 Hz); the respective nominal mains frequency then applies as the measuring point for common mode rejection or crosstalk.
6) Calculated value according to the equation for estimating usability at the given ambient temperature point or estimating usability over the specified ambient temperature range in operation (Tambient) as stated in chapter "General information on measuring accuracy/measurement uncertainty". In real use, for example at a relatively constant ambient temperature Tambient, a lower (better) achievable uncertainty is attained. A specific calculation according to chapter "General information on measuring accuracy/measurement uncertainty" is recommended, especially if the instrument allows a wider ambient temperature range in operation than 0...55 °C.
Fig.10: Representation 0…10 V measurement range|
Note: The channel also works in electrically bipolar mode and records negative values in the unipolar measuring ranges (measurement from 0 V, 0 mA, 4 mA, 0 Ω). This enables the channel to provide a precise diagnosis even with signals < 0. In these measuring ranges the limit value for the "Underrange Error" in Extended Mode is -1% of the full scale value (FSV). The limit value can be set in CoE object 0x80n0:32. This avoids irritating error messages if the channel is not wired (e.g. without sensor) or the electrical signal fluctuates slightly around zero. The process data value of 0x00000000 is not undershot. If the "UnderrangeError" detection is to be set even less sensitive, the magnitude of the negative limit value in the CoE object referred to above can be set even higher. |
