Measurement SG 1/2 bridge (half bridge) 3/5-wire connection
To determine the measuring error:
The nominal/technical measuring range is specified in "mV/V"; the maximum permitted supply voltage is 5 V. The maximum nominal measuring range that can be used for the bridge voltage is therefore ±16 mV/V ⋅ 5 V = ±80 mV; the internal circuits are designed for the 160 mV of the full bridge measurement.
The internal measurement is ratiometric, i.e. the feed voltage and the bridge voltage are not measured absolutely, but as a ratio.
The integrated supply can be used as power supply. An external supply is permitted, as long as 5 V is not exceeded.
The following is the specification given for the 5 wire connection. External line resistances are compensated by the 5 wire connection and the half bridge is detected directly from the measuring channel.
In the 3 wire connection, the terminal generally has the same specification, but its view of the connected half bridge is clouded by the unclear and temperature-dependent lead resistances within cables and connectors. In this respect, the overall system "half bridge + leads + measurement channel" will practically not achieve specification values given below.
The lead resistances (cables, connectors, ...) have an effect especially on the gain error, also depending on the temperature. The gain error can be estimated by:
(R+uv (1 + ∆T ⋅ TcCu) + R-uv(1 + ∆T ⋅ TcCu) )/Rnom with TcCu~3930 ppm/K, Rnom
e.g. 350 Ω and R+uv or R-uv lead resistances respectively.
The use of the measurement channel in the 5 wire connection is recommended.
To calculate the R1/2 half bridge:
Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 1:
R3/4 are the internal switchable supplementary resistors of the terminal. They have a high resistance of a few kΩ compared to R1/2 and thus do not significantly load the internal supply.
Other half-bridge configurations (e.g. R1/4 or R1/3 variable) cannot be connected.
The strain relationship (µStrain, µε) is as follows:
Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 2:
N should be chosen based on the mechanical configuration of the variable resistors (Poisson, 2 active uniaxial, …). The channel value (PDO) is interpreted directly [mV/V].
Common data
Measurement mode
StrainGauge/SG 1/2-Bridge 3/5-wire
16 mV/V
2 mV/V
Integrated power supply
1…5V adjustable, max. supply/Excitation 21 mA (internal electronic overload protection) therefore
  • 120R SG: up to 2.5 V
  • 350R SG: up to 5.0 V
Measuring range, nominal
-16 … 16 mV/V
-2 … 2 mV/V
Measuring range, end value (full scale value)
16 mV/V
2 mV/V
Measuring range, technically usable
-17.179 … 17.179 mV/V
-2.147 … 2.147 mV/V
PDO resolution
24 bit (including sign)
PDO LSB (Extended Range)
0.128 ppm
PDO LSB (Legacy Range)
0.119… ppm
Note: specifications apply for 3.5 V SG excitation and symmetric 350R SG.
Note: adjustment of the half-bridge measurement and thus validity of the data from production week 2018/ 50
Specific data ELM3702-0101 (preliminary data in cursive format)
Measurement mode
Measuring bridge/SG/StrainGauge/
16 mV/V
2 mV/V
Basic accuracy: Measuring deviation at 23°C, with averaging, typ. 2) 6)
without Offset
< ±0.0145 %FSV
< ±145 ppmFSV
< ±2.32 µV/V
< ±0.09 %FSV
< ±900 ppmFSV
< ±tbd.
incl. Offset
< ±0.041 %FSV
< ±410 ppmFSV
< ±6.56 µV/V
< ±0.27 %FSV
< ±2700 ppmFSV
< ±tbd.
Extended basic accuracy: Measuring deviation at 0…55°C, with averaging, typ. 2) 6)
without Offset
< ±0.053 %FSV
< ±530 ppmFSV
< ±8.48 µV/V
< ±tbd. %FSV
< ±tbd. ppmFSV
< ±tbd.
incl. Offset
< ±0.0655 %FSV
< ±655 ppmFSV
< ±10.48 µV/V
< ±tbd. %FSV
< ±tbd. ppmFSV
< ±tbd.
Offset/Zero point deviation (at 23°C)
EOffset
< 385 ppmFSV
< 2550 ppmFSV
Gain/scale/amplification deviation (at 23°C)
EGain
< 80 ppm
< 500 ppm
Non-linearity over the whole measuring range
ELin
< 120 ppmFSV
< 740 ppmFSV
Repeatability, over 24 h, with averaging
ERep
< 20 ppmFSV
< 120 ppmFSV
Temperature coefficient, typ.
TcGain
< 5 ppm/K
< tbd. ppm/K
TcOffset
< 15 ppmFSV/K
< 0.24 µV/V/K
< tbd. ppmFSV/K
< tbd.
Common-mode rejection ratio (without filter) 3)
DC:
tbd. Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 3: typ.
tbd. Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 4: typ.
50/60 Hz:
tbd. Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 5: typ.
tbd. Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 6: typ.
1 kHz:
tbd. Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 7: typ.
tbd. Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 8: typ.
Common-mode rejection ratio (with 50/60 Hz FIR filter) 1) 3)
DC:
tbd. Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 9: typ.
tbd. Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 10: typ.
50/60 Hz:
tbd. Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 11: typ.
tbd. Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 12: typ.
1 kHz:
tbd. Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 13: typ.
tbd. Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 14: typ.
Noise (without filtering, at 23°C)
ENoise, PtP
< 500 ppm FSV
< 3906 digits
< 8.00 µV/V
< 4000 ppm FSV
< 31250 digits
< 8.00 µV/V
ENoise, RMS
< 85 ppm FSV
< 664 digits
< 1.36 µV/V
< 660 ppm FSV
< 5156 digits
< 1.32 µV/V
Max. SNR
> 81.4 dB
> 63.6 dB
Noisedensity@1kHz
< 19.23 Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 15:
< 18.67 Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 16:
Noise (with 50/60 Hz FIR filter, at 23°C) 1)
ENoise, PtP
< 35 ppm FSV
< 273 digits
< 0.56 µV/V
< 280 ppm FSV
< 2188 digits
< 0.56 µV/V
ENoise, RMS
< 6.0 ppm FSV
< 47 digits
< 0.10 µV/V
< 46.0 ppm FSV
< 359 digits
< 0.09 µV/V
Max. SNR
> 104.4 dB
> 86.7 dB
Largest short-term deviation during a specified electrical interference test
tbd.
tbd.
Input impedance ±Input 1
(internal resistance)
Differential typ. tbd.
CommonMode typ. tbd.
Input impedance ±Input 2
3-wire:
No usage of this input in this mode
3-wire:
1) the 50/60 Hz filter setting must be selected via the settings in the CoE depending on the local mains frequency (50 or 60 Hz); the respective nominal mains frequency then applies as the measuring point for common mode rejection or crosstalk.
2) In real bridge measurement, an offset adjustment is usually carried out after installation. The given offset specification of the terminal is therefore practically irrelevant. Therefore, specification values with and without offset are given here. In practice, the offset component can be eliminated by the functions ELM Features and also ELM Features of the terminal or in the controller by a higher-level tare function. The offset deviation of a bridge measurement over time can change, therefore Beckhoff recommends a regular offset adjustment or careful observation of the change.
3) Values related to a common mode interference between SGND and internal ground.
6) Calculated value according to the equation for estimating usability at the given ambient temperature point or estimating usability over the specified ambient temperature range in operation (Tambient) as stated in chapter "General information on measuring accuracy/measurement uncertainty". In real use, for example at a relatively constant ambient temperature Tambient, a lower (better) achievable uncertainty is attained. A specific calculation according to chapter "General information on measuring accuracy/measurement uncertainty" is recommended, especially if the instrument allows a wider ambient temperature range in operation than 0...55 °C.
Notice
Transition resistances of the connection contacts
The transition resistance values of the connection contacts affect the measurement. The measuring accuracy can be further increased by a user-side adjustment with the signal connection plugged in.
Measurement SG 1/2 bridge (half bridge) 3/5-wire connection 17:
Validity of property values
The resistor of the bridge is positioned parallel to the internal resistor of the terminal and leads to an offset shifting respectively. The Beckhoff factory calibration will be carried out with the half bridge 350 Ω, thus the values specified above are directly valid for the 350 Ω half bridge. By connection of another dimensioned half-bridge is to:
    • perform a balancing (offset correction) by the terminal itself or the control/PLC on application side
    • or the abstract offset error have to be entered into the balancing parameter S0 of the terminal. Example: a 350 Ω half bridge correlates by the compensated effect of the input resistor (2 MΩ) during factory calibration 0.26545 %FSV (16 mV/V), that corresponds to 20738 digits.