Measurement SG 1/2 bridge (half bridge) 3/5-wire connection
To determine the measuring error:
The nominal/technical measuring range is specified in "mV/V"; the maximum permitted supply voltage is 5 V.
The maximum nominal measuring range that can be used for the bridge voltage is therefore
±16 mV/V ⋅ 5V = ±80 mV the internal circuits are designed for the 160 mV of the full bridge measurement.
The internal measurement is ratiometric, i.e. the feed voltage and the bridge voltage are not measured absolutely, but as a ratio.
The integrated supply can be used as power supply. An external supply is permitted, as long as 5 V is not exceeded.
The transition resistance values of the terminal contacts affect the measurement. The measuring error can be reduced further through compensation by the user, with active signal connection.
![]() | Validity of property values The resistor of the bridge is positioned parallel to the internal resistor of the terminal and leads to an offset shifting respectively. The Beckhoff factory calibration will be carried out with the half bridge 350 Ω, thus the values specified above are directly valid for the 350 Ω half bridge. By connection of another dimensioned half-bridge is to:
Half bridge calculation:
R3/4 are the internal switchable input resistors of the terminal. Other configurations (e.g. R1/4 or R1/3 variable) of half bridges are not supported
The strain relationship (µStrain, µε) is as follows:
N should be chosen based on the mechanical configuration of the variable resistors (Poisson, 2 active uniaxial, …).
Note: specifications apply for 2.5 V SG excitation.
Measurement mode | SG 1/2 bridge | |
Measuring range, nominal | -16…+16 mV/V [corresponds to ±16000 µε at K=2, N=2] | |
Measuring range, end value (full scale value) | 16 mV/V | |
Measuring range, technically usable | -17.179…+17.179 mV/V | |
PDO resolution | 24 bit (including sign) | |
PDO LSB (Extended Range) | 2.048 nV/V | |
PDO LSB (Legacy Range) | 1.907.. nV/V | |
Basic accuracy: Measuring deviation at 23°C, with averaging (2 | 3-wire connection: ±0.27% typ. (relative to full scale value, q.v. note) 5-wire connection: ±0.09% typ. (relative to full scale value, q.v. note) | |
Integrated power supply | 0.5…5 V adjustable, max. supply/excitation 21 mA (internal electronic overload protection) therefore
Temperature coefficient | TcTerminal | <75 ppm/K typ. (rel. to full scale value, since production 2016 week 10; refer to serial number) <250 ppm/K typ. (rel. to full scale value, production until 2016 week 09; refer to serial number) |
Input impedance ±Input 1 (Internal resistance) | Differential 4.1 MΩ || 11 nF typ. | |
CommonMode 1 MΩ || 40 nF typ. Methodology: Resistor against -UV, capacitance against SGND | ||
Input impedance ±Input 2 (Internal resistance) | Differential 5 MΩ || 10 nF typ. | |
CommonMode 1.25 MΩ || 40 nF typ. Methodology: Resistor against -UV (+2.5V), capacitance against SGND | ||
2) Dominant part of the basic accuracy is its offset specification (see following tables). The process described in the Offset correction section eliminates this proportion and increases the measurement accuracy considerably.
Measurement mode | SG 1/2 bridge (3 wire) | ||||
Offset/Zero Point deviation (at 23°C) | EOffset | < 2650 ppmFSV (valid since 2018 week 01; | |||
Gain/scale/amplification deviation (at 23°C) | EGain | < 290 ppm | |||
Non-linearity over the whole measuring range | ELin | < 390 ppmFSV | |||
Repeatability | ERep | < 50 ppmFSV | |||
Noise (without filtering)4 | ENoise. PtP | < 950 ppmFSV | < 7422 [digits] | ||
ENoise. RMS | < 145 ppmFSV | < 1133 [digits] | |||
Max. SNR | > 76.8 dB | ||||
Noisedensity @1kHz | < 32.81 | ||||
Noise (with 50 Hz FIR filtering)4 | ENoise. PtP | < 70 ppmFSV | < 547 [digits] | ||
ENoise. RMS | < 15 ppmFSV | < 117 [digits] | |||
Max. SNR | > 96.5 dB | ||||
Common-mode rejection ratio (without filtering)3 | DC: | 50 Hz: | 1 kHz: | ||
Common-mode rejection ratio (with 50 Hz FIR filtering)3 | DC: | 50 Hz: | 1 kHz: | ||
Largest short-term deviation during a specified electrical interference test | ±0.2%FSV = ±2000 ppmFSV typ. | ||||
3) Values related to a common mode interference between SGND and internal ground.
4) Specifications valid for HW Version ≥ 10, only! Valid specifications until HW Version 10 are as follows:
Measurement mode | SG 1/2 bridge (3 wire), < HW10 | ||
Noise (without filtering) | ENoise. PtP | < 1100 ppmFSV | < 8594 [digits] |
ENoise. RMS | < 160 ppmFSV | < 1250 [digits] | |
Max. SNR | > 75.9 dB | ||
Noisedensity @1kHz | < 36.2 | ||
Noise (with 50 Hz FIR filtering) | ENoise. PtP | < 120 ppmFSV | < 938 [digits] |
ENoise. RMS | < 19 ppmFSV | < 148 [digits] | |
Max. SNR | > 94.4 dB | ||
Measurement mode | SG 1/2 bridge (5 wire) | ||||
Offset/Zero Point deviation (at 23°C) | EOffset | < 750 ppmFSV (valid since 2018 week 01; | |||
Noise (without filtering)4 | ENoise. PtP | < 1200 ppmFSV | < 9375 [digits] | ||
ENoise. RMS | < 175 ppmFSV | < 1367 [digits] | |||
Max. SNR | > 75.1 dB | ||||
Noisedensity @1kHz | < 39.6 | ||||
ENoise. RMS | < 14 ppmFSV | < 109 [digits] | |||
Max. SNR | > 97.1 dB | ||||
Measurement mode | SG 1/2 bridge (5 wire), < HW10 | ||
Noise (without filtering) | ENoise. PtP | < 1600 ppmFSV | < 12500 [digits] |
ENoise. RMS | < 240 ppmFSV | < 1875 [digits] | |
Max. SNR | > 72.4 dB | ||
Noisedensity @1kHz | < 54.31 | ||
Noise (with 50 Hz FIR filtering) | ENoise. PtP | < 96 ppmFSV | < 750 [digits] |
ENoise. RMS | < 16 ppmFSV | < 125 [digits] | |
Max. SNR | > 95.9 dB | ||
The channel value (PDO) is interpreted directly [mV/V]:
Fig.45: Representation measuring range SG 1/2 bridge
typ.
typ.
