General technical data
Analog inputs
EL3314-0002
Number of inputs
4
Thermocouple sensor types, measured variables
Type B, C, E, J, K, L, N, R, S, T, U (default setting: type K), voltage measurement
Connection technology
2-wire
Maximum cable length to the thermocouple
30 m (without protective measures), suitable surge protection must be provided for longer cable lengths
Resolution
internal 24 bit
Sampling type
simultaneous
Ground reference
Differential
Conversion time
approx. 2.4 s to 4 ms depending on configuration and filter setting;
Preset: approx. 120 ms at 50/60 Hz
Input filter cut-off frequency
typ. 1 kHz
Software filter
2.5...4000 Hz, adjustable, notch characteristic; preset 50/60 Hz
additional low-pass filter possible
Wire break detection
yes (can be disabled)
Supports NoCoeStorage function
yes
Special features
high-precision, electrical isolation
Voltage measurement
EL3314-0002
Measuring ranges
± 78 mV
± 2.5 V
Resolution
1 µV per digit
Measurement uncertainty
See Measurement ±78 mV...±2.5 V
Temperature measurement
EL3314-0002
Electrical measuring range used
± 78 mV
Measuring ranges
Type B: +200...+1820 °C
Type C: 0...+2320 °C
Type E: -270...+1000 °C
Type J: -210...+1200 °C
Type K: -270…+1372 °C (preset)
Type L: -50...+900 °C
Type N: -270...+1300 °C
Type R: -50...+1768 °C
Type S: -50...+1768 °C
Type T: -270...+400 °C
Type U: -50...+600 °C
Resolution
Temperature display 0.1/0.01/0.001 °C per digit, preset 0.01 °C
See Thermocouples measurement
Supply and potentials
EL3314-0002
Power supply for the electronics
via the E-bus
Current consumption via E-bus
typ. 200 mA
Electrical isolation
2.5 kV functional isolation (test voltage 7 s channel/channel and channel/fieldbus, production test)
Max. potential ±TC to ground
2.5 kV (test voltage production test)
Max. differential voltage between the ±TC inputs
±15 V continuous
Communication
EL3314-0002
Configuration
via TwinCAT System Manager
Width in the process image
max. 24 bytes input,
max. 8 bytes output
Distributed Clocks
Environmental conditions
EL3314-0002
Permissible temperature range during operation
0 °C... + 55 °C
Permissible temperature range during storage
-25 °C... + 85 °C
Permissible relative air humidity
95 %, no condensation
General data
EL3314-0002
Dimensions (W x H x D)
approx. 15 mm x 100 mm x 70 mm (width aligned: 12 mm)
Weight
approx. 60 g
Installation
on 35 mm mounting rail, conforms to EN 60715
Installation position
To ensure enhanced measuring accuracy, the terminal must be installed in the prescribed standard position! See note!
Standards and approvals
EL3314-0002
Protection rating
IP20
Vibration / shock resistance
conforms to EN 60068-2-6 / EN 60068-2-27
EMC immunity / emission
conforms to EN 61000-6-2 / EN 61000-6-4
Identification / approval*)
CE, UKCA, EAC
cULus
Terminal
Extended features
TwinSAFE SC
high-precision
Calibration certificate
EL3314
No
No
No
No
EL3314-0002
Yes
No
Yes
No
EL3314-0010
No
No
Yes
No
EL3314-0030
No
No
Yes
External calibration certificate (ISO17025 or DAkkS certificate)
EL3314-0090
No
Yes
No
No
EL3314-0092
Yes
Yes
Yes
No