General information on commissioning the measuring bridge

For each channel of the measuring bridge, the measured value is calculated using the quotient of UBridge and USense for the respective channel and specified as an mV/V value. The relevant CoE indices n for the respective analog inputs per channel are listed below. This information applies to CoE objects that are described as 0x80n0.

The calculated measured values of the measuring bridge are made available as a standard configuration once they have been scanned in TwinCAT or manually added to the configuration in the PDO as a Real32 value.

Channel 1

Calculation via

PDO data type (default)

CoE index n

RMB

UBridge Ch. 1 / USense Ch. 1

REAL32

2
3 (for sensor-specific settings)

Channel 2*)

Calculation via

PDO data type (default)

CoE index n

RMB

UBridge Ch. 2 / USense Ch. 2

REAL32

6
7 (for sensor-specific settings)

*) only valid for ED3362-0x00

This means that only the default settings can be used for quick commissioning.

  1. Choose the Predefined PDO “RMB Standard (Real32)” (default)
  2. Specify the supply voltage (default: 10 V)
  3. Set the measuring range for UBridge (default: ±20 mV)
  4. Set the measuring range forUSense (default: ±12 V)
  5. Perform sensor calibration
  6. Read the measured value as a real value in mV/V

Optimum signal quality is achieved when using twisted-pair and shielded cables. A twisted pair should be used to connect the strain gauge power supply UExc, the sense lines USense, and the bridge differential voltage UBridge.

Notice

Short-circuiting unused inputs

Unused inputs on the same device should be properly short-circuited directly at the terminal points using short lines to prevent crosstalk!

Bridge supply UExc

The product is designed for 4-wire or 6-wire connection. The measuring bridge can be powered by the internally generated supply voltage UExc or externally. By feeding the bridge supply voltage UExc back to the measuring point USense, the influence of line losses, temperature, and drift effects is minimized.

The sensor power supply UExc is generated in the terminal from the 24 V of the power contacts and can be switched between 5 V, 10 V, and OFF [external supply required].

The sensor power supply output at the UExc terminal points is not used internally. If a sensor is used that has no additional connections for supply voltage feedback in addition to connections for the UExc supply and connections for the UBridge bridge voltage, wire jumpers must be placed between the UExc and USense terminal points in order to measure the supply voltage directly at the terminal.

Alternatively, the measuring bridge can be supplied by an external source. This means that the terminal can be operated in a 4-wire connection, as only the bridge voltage UBridge and the feedback of the external supply voltage USense are connected to the terminal and measured. If the bridge is supplied externally, the maximum permissible supply voltage is 12 V, as this is the full scale value for measuring the reference voltage USense.

Parallel connection of strain gauges

Parallel operation of load cells is permitted. Several sensors can be connected in parallel per channel, but the maximum output power of the supply must be taken into account, as the effective bridge resistance of the connected strain gauge load cells is significantly reduced by the parallel connection and power consumption is thus increased.

Each terminal can supply a current of up to 100 mA. This means that, for example, up to 7 load cells with a bridge resistance of 350 Ω each can be connected to a channel in parallel with a 5 V supply. A maximum of 3 load cells can be connected for a 10 V supply.

This specific calculation applies for internal supply and must be adjusted accordingly for external supply.

Load cells approved and calibrated for parallel operation by the load cell manufacturer should be used. The nominal characteristic values [mV/V], zero offset [mV/V], and impedance [Ω, ohm] are then usually adjusted accordingly.

General information on commissioning the measuring bridge 1:
Fig.21: Parallel connection of strain gauges