Sensor calibration
The sensor calibration adjusts the device to the characteristic curve of the sensor element. Two values are required for this procedure: the output value without a load (‘zero balance’) and fully loaded (‘rated output’). These values can be determined by a calibration protocol or by a calibration using calibration weights.
Sensor calibration 1:
Fig.44: Adaptation to the sensor curve
Sensor calibration 2:
Calibration
Calibration is of great importance for the accuracy of the system. In order to increase this, the filter should be set as high as possible over the entire calibration phase. It may take several seconds before a static value is obtained.
Sensor calibration 3:
Local storage
The values modified during the theoretical and practical calibration are stored in a local EEPROM. This can be written to up to 1 million times. In order to prolong the service life of the EEPROM, therefore, the commands should not be executed cyclically.
Adjustment in the system
In the “practical” calibration, measurement is performed with the scales unloaded first, then with a defined load on the scales. The device automatically calculates the existing sensor characteristic values from the measured values.
To do this, follow the procedure below. The adjustment is channel-specific (n = 3 (Ch. 1), n = 7 (Ch. 2).
  1. Perform CoE reset with object 0x1011:01 (see Restoring the delivery state)
  2. Set scale factor(0x80n0:27) to the desired value (default: 1000, 1 kg is displayed as 1000 g)
  3. Set gravity of earth (0x80n0:26) if necessary (default: 9.80665016)
  4. Set filter 1 (0x80n0:15) to the highest level: IIR8
  5. Specify the nominal load of the sensor in 0x80n0:24 (“Nominal load”). The unit or scaling must correspond to the set scale factor, (0x80n0:27). For example, if the nominal load is specified as 4 kg and the scale factor = 1000 (output in grams), 4000 must be entered as the nominal load.
  6. Zero adjustment: Do not load the scales
    • Wait until the measured value shows an unchanging value for at least 10 seconds (Control.PDO Steady state = 1)
    • Then execute the command “0x0101” (257dec) on CoE object 0xB0n0:01. This command enters the current measured value into the ‘Zero balance’ object.
    • Check: CoE objects 0xB0n0:02 and 0xB0n0:03 must contain “0” after execution
  7. Gain adjustment: Apply a reference load to the balance.
    • The reference weight should be at least 20% of the nominal load. The larger the reference load, the better the sensor values can be calculated.
    • In object 0x80n0:28 (“Reference load”), enter the load in the same unit as the nominal load (0x80n0:24).
    • Then execute the command “0x0102” (258dec) on CoE 0xB0n0:01. This command enters the current measured value in the “Sensitivity” object.
  8. Reset: Execute command “0x0000” (0dec) on CoE object 0xB0n0:01.
  9. Set the filter to a lower stage.
Theoretical adjustment according to sensor adjustment protocol
The sensor characteristic values according to the manufacturer's certificate are communicated directly to the device here so that it can calculate the load.
  2. Set gravity of earth (0x80n0:26) (default: 9.80665016)
  3. Adopt the “Zero balance” (0x80n0:25) from the calibration protocol
  4. Transfer “Sensitivity” (mV/V value 0x80n0:23) from the calibration protocol
  6. If required, change the scaling in 0x80n0:27 (“Scale Factor”; default: 1000) to scale in grams, kilograms, tons, etc