Plausibility check of the output
After a successful run, the program outputs a number of information items:
- estimates for single poses both with automatic and manual measurements
- In addition, the single positions measured for an automatic measurement are output.
For example, the block with the calculation results for the 1st pose is:
Example
Output of the calculation results of a sphere
"Centre of sphere" is the centre of the calibration sphere in XYZ coordinates for the pose specified. Radius “R” specifies the radius of the calibration sphere (more exactly: the sum of the radii of the calibration sphere and the dial gauge tip or switching point of the probe tip).
Criterion (A) – radii of the calibration spheres
The calibration sphere radii measured across all poses should be constant except for the measurement precision. If this is not the case, the calibration sphere was not correctly measured in one or several poses. Repeat the measurement for these poses.
Criterion (B) – Residuals
The residuals (Residual[i]) should be within the range of a few 1/100mm. They indicate the amount by which each of the measurement points deviate from the calibration sphere surface. If a measurement point deviates strongly from the sphere surface, reject the affected pose and repeat this measurement.
Calculation output
The 2nd part of the output contains the calculated **Correction values for kinematics P-CHAN-00438 and several additional items of information. This block looks like this, for example for kinematic 80:
Example
Output of calculated kinematic parameters
Besides the time stamp, the kinematic ID and the number of poses, the solution precision is also output, among other items. This value describes in mm the largest deviation between model and measurement values.
The deviation is produced by measurement errors or by errors not mapped in the model, e.g. axis misalignments, etc.
![]() | The correction parameters are output in encrypted form (P-CHAN-00438)and automatically decrypted later when they are read. |
Criterion (C) – solution precision
The solution precision should not exceed a certain limit. A limit for this value can be specified in the main program. If this limit is exceeded, an error is output and the measurement equipment, measurement method etc. must be checked.
- Measurement equipment: The measurement equipment must measure the shape of the calibration sphere correctly from all directions. The measurement precision obtained must be as small as possible.
- Measurement method: The measurement must be taken as perpendicularly to the surface of the calibration sphere as possible to avoid distortion of the contour measured. To ensure that this occurs with the automatic measurement, kinematic parameters which roughly describe the kinematic should be entered previously.
- Machine errors: Rotary axis zero positions, the rectangularity of linear axes, the positioning and repeatable precision of all axes and position-dependent errors of the rotary axes
Criterion (D) – kinematic parameters calculated
If the check is successful, the above mentioned block can be copied to the channel parameter list. After the latest import of the channel parameters, the compensation values are available to the kinematic.
Notice | |
Then a careful, slow test run should be executed with the kinematic enabled to check whether the compensation values have the correct effectiveness. This also checks that the transfer of compensation values is error-free and that no unintended collisions occur. |