General technical data
All values are typical values over the entire temperature range, unless stated otherwise.
EtherCAT P
Connection
2 x M8 socket, 4-pin, P-coded, red
Supply voltages
Connection
See EtherCAT P connection
US nominal voltage
24 VDC (-15 % / +20 %)
US sum current: IS,sum
max. 3 A
Current consumption from US
100 mA
Rated voltage UP
UP sum current: IP,sum
max. 3 A
Current consumption from UP
None. UP is only forwarded.
Thermocouple inputs
Number
4
Connector
4 x M12 socket
Cable length to thermocouple
max. 30 m
Sensor types
  • Thermocouples
  • Sensors with voltage output up to ±75 mV
Electrical isolation
The measuring channels have a common isolated ground potential.
Measuring ranges
Thermocouples: depending on the thermocouple type.
Voltage measurement: ±30 mV, ±60 mV, ±75 mV
Measurement uncertainty
see chapter Thermocouples measurement.
Digital resolution
16 bits
Value of an LSB
Adjustable for thermocouple measurement:
  • 0.1 °C (factory setting)
  • 0.01 °C
For voltage measurement:
  • Measuring range 30 mV: 1 µV
  • Measuring range 60 mV: 2 µV
  • Measuring range 75 mV: 4 µV
Filter
Digital filter. Filter frequency adjustable from 5 Hz … 30 kHz
Conversion time
approx. 2.5 s to 20 ms, depending on configuration and filter setting.
Default: approx. 250 ms
Diagnosis
  • Open-circuit recognition
  • Limit value monitoring
Housing data
Dimensions W x H x D
30 mm x 126 mm x 26.5 mm (without connectors)
Weight
approx. 165 g
Installation position
variable
Material
PA6 (polyamide)
Environmental conditions
Ambient temperature during operation
-25 … +60 °C
-25 … +55 °C according to cULus
Ambient temperature during storage
-40 … +85 °C
Vibration resistance, shock resistance
conforms to EN 60068-2-6 / EN 60068-2-27
Additional checks
EMC immunity / emission
conforms to EN 61000-6-2 / EN 61000-6-4
Protection class
IP65, IP66, IP67 (conforms to EN 60529)
Approvals / markings
Approvals / markings *)
CE, cULus
Additional tests
The devices have undergone the following additional tests:
Test
Explanation
Vibration
10 frequency sweeps in 3 axes
5 Hz < f < 60 Hz displacement 0.35 mm, constant amplitude
60.1 Hz < f < 500 Hz acceleration 5 g, constant amplitude
Shocks
1000 shocks in each direction, in 3 axes
35 g, 11 ms
Overview of suitable thermocouples
The following thermocouple types are suitable for temperature measurement:
Type (conforms to EN60584-1)
Element
Implemented temperature range
Color coding (sheath - positive pole - negative pole)
B
Pt30%Rh-Pt6Rh
200 °C to 1820 °C
grey - grey - white
C *
 W5%Re-W25%Re
0 °C to 2320 °C
n.d.
E
NiCr-CuNi
-100°C to 1000°C
violet - violet - white
J
Fe-CuNi
-100°C to 1200°C
black - black - white
K
NiCr-Ni
-200°C to 1372°C
green - green - white
L **
Fe-CuNi
0 °C to 900 °C
blue - red - blue
N
NiCrSi-NiSi
-100°C to 1300°C
pink - pink - white
R
Pt13%Rh-Pt
-50°C to 1767°C
orange - orange - white
S
Pt10%Rh-Pt
-50°C to 1760°C
T
Cu-CuNi
-200°C to 400°C
brown - brown - white
U **
Cu-CuNi
0 °C to 600 °C
brown - red - brown
*not standardized according to EN60584-1
** according to DIN 43710
