Technical data
All values are typical values over the entire temperature range, unless stated otherwise.
Supply voltages | |
Connection | Input: M8 connector, 4-pin, A-coded Downstream connection: M8 socket, 4-pin, A-coded |
US nominal voltage |
24 VDC (-15 % / +20 %) |
US sum current: IS,sum | max. 4 A |
Current consumption from US |
120 mA |
Rated voltage UP | 24 VDC (-15 % / +20 %) |
UP sum current: IP,sum | max. 4 A |
Current consumption from UP | None. UP is only forwarded. |
Thermocouple inputs | |
Number | 4 |
Connector | 4x M12 socket, 5-pin, A-coded |
Cable length | max. 30 m between box and thermocouple |
Sensor types |
Electrical isolation | The measuring channels have a common isolated ground potential. |
Measuring ranges | Thermocouples: depending on the thermocouple type. Voltage measurement: ±30 mV, ±60 mV, ±75 mV |
Measurement uncertainty | see chapter Thermocouple specifications. |
Digital resolution | 16 bit |
Value of an LSB | Adjustable for thermocouple measurement:
For voltage measurement:
Filter | Digital filter. Filter frequency adjustable from 5 Hz … 30 kHz |
Conversion time | approx. 2.5 s to 20 ms, depending on configuration and filter setting. Default: approx. 250 ms |
Diagnosis |
1) This value corresponds to the current carrying capacity of the connections for the supply voltages.
Housing data | |
Dimensions W x H x D | 30 mm x 126 mm x 26.5 mm (without connectors) |
Weight | approx. 165 g |
Material |
PA6 (polyamide) |
Environmental conditions | |
Ambient temperature during operation | -25 … +60 °C |
Ambient temperature during storage | -40 … +85 °C |
Vibration resistance, shock resistance | conforms to EN 60068-2-6 / EN 60068-2-27 |
Protection rating | IP65, IP66, IP67 (according to EN 60529) |
Approvals / markings | |
Approvals / markings *) |
Additional tests
The devices have undergone the following additional tests:
Test | Explanation |
Vibration | 10 frequency sweeps in 3 axes |
5 Hz < f < 60 Hz displacement 0.35 mm, constant amplitude | |
60.1 Hz < f < 500 Hz acceleration 5 g, constant amplitude | |
Shocks | 1000 shocks in each direction, in 3 axes |
35 g, 11 ms |
Overview of suitable thermocouples
The following thermocouple types are suitable for temperature measurement:
Type (conforms to EN60584-1) | Element | Implemented temperature range | Color coding (sheath - positive pole - negative pole) |
B | Pt30%Rh-Pt6Rh | 200 °C to 1820 °C | grey - grey - white |
C * | W5%Re-W25%Re | 0 °C to 2320 °C | n.d. |
E | NiCr-CuNi | -100°C to 1000°C | violet - violet - white |
J | Fe-CuNi | -100°C to 1200°C | black - black - white |
K | NiCr-Ni | -200°C to 1372°C | green - green - white |
L ** | Fe-CuNi | 0 °C to 900 °C | blue - red - blue |
N | NiCrSi-NiSi | -100°C to 1300°C | pink - pink - white |
R | Pt13%Rh-Pt | -50°C to 1767°C | orange - orange - white |
S | Pt10%Rh-Pt | -50°C to 1760°C | orange - orange - white |
T | Cu-CuNi | -200°C to 400°C | brown - brown - white |
U ** | Cu-CuNi | 0 °C to 600 °C | brown - red - brown |
*not standardized according to EN60584-1
** according to DIN 43710