Measurement IEPE ±40 mV
Measurement mode |
±40 mV | ||
Measuring range, nominal |
-40…+40 mV | ||
Measuring range, end value (FSV) |
40 mV | ||
Measuring range, technically usable |
-42.95…+42.95 mV | ||
PDO resolution (including sign) |
24 bit |
16 bit 2) | |
PDO LSB (Extended Range) |
5.12 nV |
1.31072 µV | |
PDO LSB (Legacy Range) |
4.768.. nV |
1.220.. µV | |
Input impedance ±Input 1 (internal resistance) |
Differential typ. 2 MΩ || 1 nF CommonMode typ. 10 nF against SGND |
2) The analog measurement is always done with 24 bits, in 16-bit mode the eight least significant bits are cut off. For further information see chapter “Commissioning”/ “Process data overview”
Preliminary specifications:
Measurement mode |
±40 mV | ||||
Basic accuracy: Measuring deviation at 23°C, with averaging |
< ±0.02% = 200 ppmFSV typ. | ||||
Offset/Zero point deviation (at 23°C) |
EOffset |
< 175 ppmFSV | |||
Gain/scale/amplification deviation (at 23°C) |
EGain |
< 65 ppm | |||
Non-linearity over the whole measuring range |
ELin |
< 45 ppmFSV | |||
Repeatability, over 24 h, with averaging |
ERep |
< 30 ppmFSV | |||
Noise (without filtering) |
ENoise, PtP |
< 375 ppmFSV |
< 2930 [digits] | ||
ENoise, RMS |
< 75 ppmFSV |
< 586 [digits] | |||
Max. SNR |
> 82.5 dB | ||||
Noisedensity@1kHz |
< 42.43
Noise (with 50 Hz FIR filter) |
ENoise, PtP |
< 40 ppmFSV |
< 313 [digits] | ||
ENoise, RMS |
< 5.5 ppmFSV |
< 43 [digits] | |||
Max. SNR |
> 105.2 dB | ||||
Temperature coefficient |
TcGain |
8 ppm/K typ. | |||
TcOffset |
6 ppmFSV/K typ. | ||||
Common-mode rejection ratio (without filter) |
DC: >115 dB typ. |
50 Hz: >105 dB typ. |
1 kHz: >80 dB typ. | ||
Common-mode rejection ratio (with 50 Hz FIR filter) |
DC: >115 dB typ. |
50 Hz: >115 dB typ. |
1 kHz: >115 dB typ. | ||
Largest short-term deviation during a specified electrical interference test |
Value to follow |
![Measurement IEPE ±40 mV 2:](Images/png/1904223499__Web.png)
Note: In Extended Range Mode the Underrange/Overrange display in the PDO status has the character of an information/warning when the nominal measuring range is exceeded, i.e. no Error is displayed in the PDO status and LED. If the technical measuring range is also exceeded, Error = TRUE is also displayed. The detection limit for Underrange/Overrange Error can be set in the CoE. In Legacy Range mode, an Underrange/Overrange event also leads to an Error in the PDO status. |
Frequency response: see specification of ±10 V measurement range
