Measurement IEPE 0…10 V
Measurement mode
0…10 V
Measuring range, nominal
0…10 V
Measuring range, end value (FSV)
10 V
Measuring range, technically usable
0…+10.737 V
PDO resolution (unsigned)
23 bit
15 bit 2)
PDO LSB (Extended Range)
1.28 µV
327.68 µV
PDO LSB (Legacy Range)
1.192.. µV
305.18.. µV
Input impedance ±Input 1
(internal resistance)
Differential typ. 2 MΩ || 1 nF
CommonMode typ. 10 nF against SGND
Measurement mode
0…10 V
Basic accuracy: Measuring deviation at 23°C, with averaging 1) 6)
< ±0.0075 % = 75 ppmFSV typ.
< ±0.8 mV typ.
Extended basic accuracy: Measuring deviation at 0…55°C, with averaging 1) 6)
< ±0.0105 %, < ±105 ppmFSV typ.
< ±1.1 mV typ.
Offset/Zero point deviation (at 23°C) 1)
EOffset
< 20 ppmFSV
Gain/scale/amplification deviation (at 23°C) 1)
EGain
< 60 ppm
Non-linearity over the whole measuring range 1)
ELin
< 40 ppmFSV
Repeatability, over 24 h, with averaging 1)
ERep
< 5 ppmFSV
Temperature coefficient 1)
TcGain
< 2 ppm/K typ.
TcOffset
< 1 ppmFSV/K typ.
< 10 µV/K typ.
Preliminary data:
Measurement mode
0…10 V
Noise (without filtering)
ENoise, PtP
< 100 ppmFSV
< 781 digits
ENoise, RMS
< 18 ppmFSV
< 141 digits
Max. SNR
> 94.9 dB
Noisedensity@1kHz
< 2.55 Measurement IEPE 0…10 V 1:
Noise (with 50/60 Hz FIR filter) 5)
ENoise, PtP
< 10 ppmFSV
< 78 digits
ENoise, RMS
< 2.0 ppmFSV
< 16 digits
Max. SNR
> 114.0 dB
Common-mode rejection ratio (without filter)
DC: >115 dB typ.
50/60 Hz: >105 dB typ.
1 kHz: >80 dB typ.
Common-mode rejection ratio (with 50/60 Hz FIR filter) 5)
DC: >115 dB typ.
50/60 Hz: >115 dB typ.
1 kHz: >115 dB typ.
Largest short-term deviation during a specified electrical interference test
±0.03 % = 300 ppmFSV typ.
1) Valid for ELM3602-00x0 from HW08, ELM360x-00x2 from HW09, ELM3604-00x0 from HW10; specifications of predecessor-HW on request
2) The analog measurement is always done with 24 bits, in 16-bit mode the eight least significant bits are cut off. For further information see chapter “Commissioning”/ “Process data overview”
5) the 50/60 Hz filter setting must be selected via the settings in the CoE depending on the local mains frequency (50 or 60 Hz); the respective nominal mains frequency then applies as the measuring point for common mode rejection or crosstalk.
6) Calculated value according to the equation for estimating usability at the given ambient temperature point or estimating usability over the specified ambient temperature range in operation (Tambient) as stated in chapter "General information on measuring accuracy/measurement uncertainty". In real use, for example at a relatively constant ambient temperature Tambient, a lower (better) achievable uncertainty is attained. A specific calculation according to chapter "General information on measuring accuracy/measurement uncertainty" is recommended, especially if the instrument allows a wider ambient temperature range in operation than 0...55 °C.
Frequency response: See data within ±10V measurement mode
Measurement IEPE 0…10 V 2:
Fig.10: Representation 0…10 V measurement range
Note: The channel also works in electrically bipolar mode and records negative values in the unipolar measuring ranges (measurement from 0 V, 0 mA, 4 mA, 0 Ω). This enables the channel to provide a precise diagnosis even with signals < 0. In these measuring ranges the limit value for the "Underrange Error" in Extended Mode is -1% of the full scale value (FSV). The limit value can be set in CoE object 0x80n0:32. This avoids irritating error messages if the channel is not wired (e.g. without sensor) or the electrical signal fluctuates slightly around zero. The process data value of 0x00000000 is not undershot.
If the "UnderrangeError" detection is to be set even less sensitive, the magnitude of the negative limit value in the CoE object referred to above can be set even higher.