Measurement ±60 V
Measurement mode | ±60 V | |
Measuring range, nominal | -60…+60 V | |
Measuring range, end value (FSV) | 60 V | |
Measuring range, technically usable | -64.414…+64.414 V | |
PDO resolution (including sign) | 24 bit | 16 bit 2) |
PDO LSB (Extended Range) | 7.68 µV | 1.966 mV |
PDO LSB (Legacy Range) | 7.152.. µV | 1.831.. mV |
Input impedance ±Input 1 (internal resistance) | Differential typ. approx. 485 kΩ || 11 nF CommonMode typ. approx. 40 nF against SGND | |
2) The analog measurement is always done with 24 bits, in 16-bit mode the eight least significant bits are cut off. For further information see chapter “Commissioning”/ “Process data overview”
Preliminary data
Measurement mode | ±60 V | |||
Basic accuracy: Measuring deviation at 23°C, with averaging | < ±(tbd) = (tbd) ppmFSV typ. | |||
Offset/Zero point deviation (at 23°C) | EOffset | < (tbd) ppmFSV | ||
Gain/scale/amplification deviation (at 23°C) | EGain | < (tbd) ppm | ||
Non-linearity over the whole measuring range | ELin | < (tbd) ppmFSV | ||
Repeatability | ERep | < (tbd) ppmFSV | ||
Temperature coefficient | TcGain | < (tbd) ppm/K typ. | ||
TcOffset | < (tbd) ppmFSV/K typ. | |||
Noise (without filtering) | ENoise, PtP | < 75 ppmFSV | < 586 [digits] | < 4.50 mV |
ENoise, RMS | < 13 ppmFSV | < 98 [digits] | < 0.75 mV | |
Max. SNR | > 98.1 dB | |||
Noisedensity@1kHz | < 10.61
Noise (with 50/60 Hz FIR filter 5)) | ENoise, PtP | < 12 ppmFSV | < 94 [digits] | < 0.72 mV |
ENoise, RMS | < 2.0 ppmFSV | < 16 [digits] | < 0.12 mV | |
Max. SNR | > 114.0 dB | |||
Common-mode rejection ratio (without filter) | DC: >(tbd) dB typ. | 50/60 Hz: >(tbd) dB typ. | 1 kHz: >(tbd) dB typ. | |
Common-mode rejection ratio (with 50/60 Hz FIR filter 5)) | DC: >(tbd) dB typ. | 50/60 Hz: >(tbd) dB typ. | 1 kHz: >(tbd) dB typ. | |
Largest short-term deviation during a specified electrical interference test | ±(tbd)% = (tbd) ppmFSV typ. | |||
5) the 50/60 Hz filter setting must be selected via the settings in the CoE depending on the local mains frequency (50 or 60 Hz); the respective nominal mains frequency then applies as the measuring point for common mode rejection or crosstalk.
Note: In Extended Range Mode the Underrange/Overrange display in the PDO status has the character of an information/warning when the nominal measuring range is exceeded, i.e. no Error is displayed in the PDO status and LED. If the technical measuring range is also exceeded, Error = TRUE is also displayed. The detection limit for Underrange/Overrange Error can be set in the CoE. In Legacy Range mode, an Underrange/Overrange event also leads to an Error in the PDO status. |