Measured value processing
The electrical measured value recorded and digitized by the ADC must or can be changed in the device in order to
Note: This CoE uses the term “calibration”, which is historically used at Beckhoff, although it has nothing to do with deviation statements of a calibration certificate. The vendor or customer calibration data/adjustment data that the device uses during operation in order to maintain the guaranteed analog accuracy is described here.
The measured value can be changed in 3 functional units, all three can be active at the same time:
Vendor Calibration
The electrical channel is calibrated by Beckhoff in the “Vendor Calibration” functional unit for compliance with the specified uncertainty (see Technical data, previously: Measuring error). The vendor calibration data from Beckhoff is available in this area.
Parameter:
Index (hex)
Data type
Meaning
80n0:0B
Enable vendor calibration
enabled by default, the data is taken into account.
80nF
Vendor calibration data
not intended for user modification
The intermediate value after this functional unit can be viewed in index 0x90n0:05 under “Vendor Calibration”.
The number of setting changes in this functional unit is counted up in index 0x90n2:11 as “Vendor Calibration Counter” and cannot be deleted.
If any parameter in the data area is changed, the counter is incremented. Further changes in the data area within the next 30 seconds are not taken into account for the counter. After this time has elapsed, a parameter change will increment the counter again.
User Calibration
The “User calibration” functional unit can be used by the user if alternative, system-dependent correction values are to be used permanently.
The gain/offset coefficients are available both in Real32 format for convenient operation and alternatively (but with the same mathematical effect) in INT16 format for compatibility with any existing code. Processing in "User Calibration" (if “Enable User calibration” = 1) is as follows:
Parameter:
Index (hex)
Data type
Meaning
80n0:0A
Enable User Calibration
disabled by default, calculation only takes place if TRUE
80n0:17
User Calibration Offset
SINT16
1 Bit = FSVnorm./32767, default: 0
80n0:18
User Calibration Gain
UINT16
1 bit corresponds to 2-16, so “1” corresponds to 0x7FFF/32767dec
80nC:01
User Calibration Data
BYTE4
4 bytes of free memory space; here it is possible to store the calibration date in the form of 8 CHAR, for example
80nC:03…0D
User Scale Gain (Real32)
REAL32
Real32 coefficients S0/S1/S2/S3/S1n of the calculation polynomial
The intermediate value after this functional unit can be viewed in index 0x90n0:06 “Value after User Calibration”.
The number of setting changes in this functional unit is counted up in index 0x90n2:12 “User Calibration Counter” (cannot be deleted).
Procedure: the counter is incremented the first time any parameter in the data area index 0x80nC or index 0x80n0:17/18 is changed; further changes in the data area in the following 30 seconds are not taken into account for the counter. After this time has elapsed, a parameter change will increment the counter again.
User Scale
The “User Scale” functional unit is intended for reinterpretations/transformations of the measured value, so “10 V” can become “50 kg” with Gain = 5. It is implemented as a linear transformation with gain/offset.
Processing in “User Scale” (if “Enable User Scale” = 1) is as follows:
“Value after User scale” = Offset + value of filter 2 * Gain
Parameters:
Index (hex)
Data type
Meaning
80n0:01
Enable User Scale
80n0:11
User Scale Offset
SINT16
is added directly in digits.
80n0:12
User Scale Gain
UINT16
80nD:1C
User Scale Offset (REAL32)
REAL32
80nD:1D
User Scale Gain (REAL32)
REAL32
The intermediate value after this functional unit can be viewed in index 0x90n0:0B “Value After User Scale”.
Measured value processing 1:
Changing the interface
When the interface is changed, the gain and offset are reset to 1 and 0 respectively!
Password protection for user data
Some user data are protected against unwanted or inadvertent writing by an additional password to be entered in CoE 0xF009:
Measured value processing 2:
Fig.43: Password protection for the 0x8000:17 and 0x8000:18 entries (example)
Use of CoE 0xF009
Password protection takes effect with the following AI settings:
Index (hex)
80n0:0A
80n0:0B
Enable Vendor Calibration
80n0:17
80n0:18
80nC
80nD:17
Low Range Error
80nD:18
High Range Error
80nD:27
Low Range Error (REAL32)
80nD:28
High Range Error (REAL32)