Technology
General description
The functionality of the EL3681 is similar to that of a commercial digital multimeter. The terminal offers the following features:
The data acquisition is shown in fig. Data flow EL3681.
Technology 1:Fig.5: Data flow EL3681
Accuracy
The unused measurement input should be connected to the COM port of the terminal in order to make the analysis as accurate as possible and minimize interference.
The measuring accuracy depends on the type of signal to be measured and on the terminal settings.
The accuracy specifications given refer to an input signal in the frequency range >0 ... 1 KHz. With higher frequencies the measuring accuracy decreases (-3 dB >500 kHz).
The accuracy values specified in the following table apply to the default settings for the terminal parameters:
- Enable vendor calibration
true
- Enable filter
true
- Frequency
50 Hz
- Zero compensation interval
Off (0)
- Presentation
Scaled (1Bit/1µV) (2)
Signal to be measured
Typ. max. deviation in % of FSV 1)
typ. temperature drift 2)6)
Measuring mode
Measuring range
40°C 3)
0 .. 55°C
ppm/°C
DC
3 V - 300 V 8)
0.01
0.2
35
300 mV 8)
0.05
0.2
35
100 mA 7)
0.1
0.5
50
1 A
0.1
0.5
50
10 A
0.2
1.2
170
AC 4) 5)
3 V - 300 V
0.25
0.75
130
300 mV
0.25
0.5
50
100 mA
0.5
1
50
1 A
0.5
0.7
50
10 A
0.5
1.2
150
Tab. 1: Measuring tolerances as a function of temperatures. FSV = full scale value.
1) In 60 Hz mode of the ADC, add 0.02 to the specified deviation
2) Values apply for a minimum terminal warm-up time of 30 min
3) Adjustment temperature is 40 °C
4) All AC voltage and AC current ranges are specified for the range from 5% to 100%
5) A non-sinusoidal AC input signal can also be measured if it does not exceed a crest factor of 3. The specified accuracy data refer to a crest factor of <= 2
6) In 60 Hz operation, an additional temperature drift of 20 ppm / °C must be expected.
7) The maximum deviation under EMC test conditions according to IEC 61131 is 1 %
8) The maximum deviation under EMC test conditions according to IEC 61131 is 0.2 %
Application note
Internal resistance
Measurement type
Measuring range
Internal resistance
DC
300 mV - 300 V
12.5 MΩ
100 mA - 1 A
0.2 Ω
10 A
3 mΩ
AC 4) 5)
300 mV - 300 V
1 MΩ, approx. 33 pF
100 mA - 1 A
0.2 Ω
10 A
3 mΩ
Table 2: Internal resistances
Fuse
Notes on replacing the fuse can be found in section Accessories.