Technical data
Technical data
EL3433
Measured values
Current, voltage, effective power, apparent power, frequency
Calculated values
Reactive power, energy, power factor (cosφ), harmonic frequencies, phase angle
Measuring voltage
max. 500 V AC 3~ (ULX-N: max. 288 V AC)
Fed-in voltages must comply with overvoltage category II
Measuring current
max. 10 A (AC) (configurable)
Input resistance voltage circuit (typ.)
1 MΩ
Input resistance current circuit (typ.)
< 3 mΩ
Fuse protection
Voltage circuit: according to the connected conductor size
Current circuit: primary side of the current transformer, according to the connected conductor size
Resolution
1 µA, 0.1 mV, 10 mW
Measuring accuracy
0.5% in relation to the full scale value (U/I) (at 0°C…55°C)
1% in relation to the full scale value (U/I) (when the extended temperature range is used)
1% calculated value (P)
Notice:
For the EL3433, an accuracy of 2% FSV (full scale value) of the largest measuring range of the terminal is valid referring to the neutral conductor current measurement. The neutral conductor current measurement is only possible for this measuring range.
Frequency range
45 Hz to 65 Hz
Signal type
any (taking into account the frequency range and the limit frequency)
Measuring procedure
True RMS calculation with 16,800 (2,800 per channel) samples/s
Sampling type
simultaneous
Ground reference
single ended
Measuring cycle time
200 ms per measured value preset, freely configurable, mains-synchronous
Electrical isolation
4500 V (connection terminal/E-bus)
Supply voltage for electronic
via the E-bus
E-Bus current consumption
200 mA typ.
Configuration
via TwinCAT System Manager
Dimensions (W x H x D)
approx. 27 mm x 100 mm x 70 mm (width aligned: 24 mm)
Weight
approx. 100 g
Mounting
on 35 mm mounting rail conforms to EN 60715
Operating temperature
-25 °C ... +60 °C (extended temperature range)
Storage temperature
-40 °C ... +85 °C
Relative humidity
95 % no condensation
Vibration/shock resistance
conforms to EN 60068-2-6 / EN 60068-2-27
EMC immunity/emission
conforms to EN 61000-6-2 / EN 61000-6-4
Protect. class / installation pos.
IP20/any
Marking / Approval*)
CE, UKCA, EAC