Thermocouples measurement
In the measuring range of a specified thermocouple type, a measured voltage is converted internally into a temperature according to the set transformation. Since the channel measures a voltage internally, the corresponding measuring error in the voltage measuring range must be used.
The following tables with the specification of the thermocouple measurement apply only when using the internal cold junction.
The EL331x-00xx can also be used with an external cold junction if required. The uncertainties must then be determined for the external cold junction on the application side. The temperature value of the external cold junction must then be communicated to the EL331x-00xx via the process data for its own calculation. The effect on the measurement of the thermocouples must then be calculated on the system side.
The specifications for the internal cold junction and the measuring range given here apply only if the following times are adhered to for thermal stabilization at constant ambient temperature:
Specification of the internal cold junction measurement
In the EL3311 each channel has its own cold junction sensor.
Measurement mode
Cold junction
Basic accuracy: Measurement deviation at 23°C, with averaging
< ±2.0°C
Temperature coefficient
Tk
< 40 mK/K
Specification - thermocouple type B
Temperature measurement thermocouple
Type B
Electrical measuring range used
± 75 mV
Measuring range, technically usable
+200°C ≈ 0.178 mV … +1820°C ≈ 13.820 mV
Measuring range, end value (full scale value)
+1820°C
Measuring range, recommended
+750°C … +1800°C
PDO LSB
0.1 / 0.01°C/digit, depending on PDO setting
Note: internally, 16 bits are used for the calculation up to the FSV; depending on the set thermocouple, therefore, jumps in value > 0.01°C occur with "resolution 0.01°C"; e.g. type B: approx. 0.05°C
Uncertainty in the recommended measuring range, with averaging
@ 23°C ambient temperature
± 8.5 K ≈ ± 0.47%FSV
@ 55°C ambient temperature
± 9.7 K ≈ ± 0.54%FSV
(Change in the measured value in relation to the change in the ambient temperature of the terminal)
Since the value is highly dependent on the sensor temperature, as can be seen in the specification plot shown below, it must basically be derived from the specification plot. For better approximation, the measurement uncertainty at Tamb= 39°C as the middle point between 23°C and 55°C is also shown informatively in order to illustrate the non-linear curve.
Measurement uncertainty for thermocouple type B:
Thermocouples measurement 1:
Specification - thermocouple type C
Type C
± 75 mV
0°C ≈ 0 mV … +2320°C ≈ 37.107 mV
+2320°C
0°C … +2000°C
PDO LSB
0.1/0.01°C/digit, depending on PDO setting
Note: internally, 16 bits are used for the calculation up to the FSV; depending on the set thermocouple, therefore, jumps in value > 0.01°C occur with "resolution 0.01°C"; e.g. type C: approx. 0.07°C
± 6.2 K ≈ ± 0.27%FSV
± 7.4 K ≈ ± 0.32%FSV
Measurement uncertainty for thermocouple type C:
Thermocouples measurement 2:
Specification - thermocouple type E
Type E
± 75 mV
-270°C ≈ -9.835 mV … +1000°C ≈ 76.373 mV
+1000°C
-100°C … +1000°C
PDO LSB
Note: internally, 16 bits are used for the calculation up to the FSV; depending on the set thermocouple, therefore, jumps in value > 0.01°C occur with "resolution 0.01°C"; e.g. type E: approx. 0.03°C
± 2.5 K ≈ ± 0.25%FSV
± 3.0 K ≈ ± 0.30%FSV
Measurement uncertainty for thermocouple type E:
Thermocouples measurement 3:
Specification - thermocouple type J
Type J
± 75 mV
-210°C ≈ -8.095 mV … +1200°C ≈ 69.553 mV
+1200°C
-100°C … +1200°C
PDO LSB
Note: internally, 16 bits are used for the calculation up to the FSV; depending on the set thermocouple, therefore, jumps in value > 0.01°C occur with "resolution 0.01°C"; e.g. type J: approx. 0.04°C
± 2.7 K ≈ ± 0.23%FSV
± 3.2 K ≈ ± 0.27%FSV
Measurement uncertainty for thermocouple type J:
Thermocouples measurement 4:
Specification - thermocouple type K
Type K
± 75 mV
-270°C ≈ -6.458 mV … +1372°C ≈ 54.886 mV
+1372°C
-100°C … +1200°C
PDO LSB
Note: internally, 16 bits are used for the calculation up to the FSV; depending on the set thermocouple, therefore, jumps in value > 0.01°C occur with "resolution 0.01°C"; e.g. type K: approx. 0.04°C
± 3.0 K ≈ ± 0.22%FSV
± 3.6 K ≈ ± 0.26%FSV
Measurement uncertainty for thermocouple type K:
Thermocouples measurement 5:
Specification - thermocouple type L
Type L
± 75 mV
-50°C ≈ -2.510 mV … +900°C ≈ 52.430 mV
+900°C
0°C … +900°C
PDO LSB
Note: internally, 16 bits are used for the calculation up to the FSV; depending on the set thermocouple, therefore, jumps in value > 0.01°C occur with "resolution 0.01°C"; e.g. type L: approx. 0.03°C
± 2.3 K ≈ ± 0.26%FSV
± 2.8 K ≈ ± 0.31%FSV
Measurement uncertainty for thermocouple type L:
Thermocouples measurement 6:
Specification - thermocouple type N
Type N
± 75 mV
-270°C ≈ -4.346 mV … +1300°C ≈ 47.513 mV
+1300°C
0 °C … +1200 °C
PDO LSB
Note: internally, 16 bits are used for the calculation up to the FSV; depending on the set thermocouple, therefore, jumps in value > 0.01°C occur with "resolution 0.01°C"; e.g. type N: approx. 0.04°C
± 3.0 K ≈ ± 0.23%FSV
± 3.5 K ≈ ± 0.27%FSV
Measurement uncertainty for thermocouple type N:
Thermocouples measurement 7:
Specification - thermocouple type R
Type R
± 75 mV
-50°C ≈ -0.226 mV … +1768°C ≈ 21.101 mV
+1768°C
+250°C … +1700°C
PDO LSB
Note: internally, 16 bits are used for the calculation up to the FSV; depending on the set thermocouple, therefore, jumps in value > 0.01°C occur with "resolution 0.01°C"; e.g. type R: approx. 0.05°C
± 6.7 K ≈ ± 0.38%FSV
± 7.7 K ≈ ± 0.44%FSV
Measurement uncertainty for thermocouple type R:
Thermocouples measurement 8:
Specification - thermocouple type S
Type S
± 75 mV
-50°C ≈ -0.236 mV … +1768°C ≈ 18.693 mV
+1768°C
+250°C … +1700°C
PDO LSB
Note: internally, 16 bits are used for the calculation up to the FSV; depending on the set thermocouple, therefore, jumps in value > 0.01°C occur with "resolution 0.01°C"; e.g. type S: approx. 0.05°C
± 7.1 K ≈ ± 0.40%FSV
± 8.1 K ≈ ± 0.46%FSV
Measurement uncertainty for thermocouple type S:
Thermocouples measurement 9:
Specification - thermocouple type T
Type T
± 75 mV
-270°C ≈ -6.258 mV … +400°C ≈ 20.872 mV
+400°C
-100°C … +400°C
PDO LSB
± 2.9 K ≈ ± 0.73%FSV
± 3.4 K ≈ ± 0.85%FSV
Measurement uncertainty for thermocouple type T:
Thermocouples measurement 10:
Specification - thermocouple type U
Type U
± 75 mV
-50°C ≈ -1.850 mV … +600°C ≈ 33.600 mV
+600°C
0°C … +600°C
PDO LSB
Note: internally, 16 bits are used for the calculation up to the FSV; depending on the set thermocouple, therefore, jumps in value > 0.01°C occur with "resolution 0.01°C"; e.g. type U: approx. 0.02°C
± 2.5 K ≈ ± 0.42%FSV
± 2.9 K ≈ ± 0.48%FSV
Measurement uncertainty for thermocouple type U:
Thermocouples measurement 11: