Technical data
All values are typical values over the entire temperature range, unless stated otherwise.
EtherCAT
Connection
2 x M8 socket, 4-pin, green
Electrical isolation
500 V
Distributed Clocks
yes
Supply voltages
Connection
Input: M8 connector, 4-pin
Downstream connection: M8 socket, 4-pin, black
US nominal voltage
24 VDC (-15 % / +20 %)
US sum current: IS,sum
max. 4 A
Current consumption from US
120 mA
+ IEPE supply current IEXCITE
Rated voltage UP
UP sum current: IP,sum
max. 4 A
Current consumption from UP
None. UP is only forwarded.
IEPE
Number of inputs
2
Connection
2 x M8 socket, four-pin, screwable, shielded
Signal voltage
IEPE constant current supply and acquisition of modulated AC voltage
Sensor state monitoring
yes, through monitoring of the bias voltage
Measuring range
preset ±5 V up to 25 kHz, ±250 mV up to 10 Hz
Input filter cut-off frequency
analog parameterizable 5th order low-pass filter up to 25 kHz, typically 0.05 Hz high-pass filter
Measurement uncertainty
< ±0.5 % (DC, relative to full scale value)
See chapter Measurement error/measurement deviation/measurement uncertainty, output uncertainty.
Resolution
16 bits (including sign)
Conversion time
20 μs (max. 50 kSamples/s)
Supply current IEXCITE
2 / 4 / 8 mA (separately adjustable for both channels) from the control voltage US
Special features
automatic anti-aliasing function, wire break detection
Housing data
Dimensions W x H x D
30 mm x 126 mm x 26.5 mm (without connectors)
Weight
approx. 165 g
Installation position
variable
Material
PA6 (polyamide)
Environmental conditions
Ambient temperature during operation
-25 … +60 °C
-25 … +55 °C according to cURus
Ambient temperature during storage
-40 … +85 °C
Vibration resistance, shock resistance
conforms to EN 60068-2-6 / EN 60068-2-27
Additional checks
EMC immunity / emission
conforms to EN 61000-6-2 / EN 61000-6-4
Protection class
IP65, IP66, IP67 (conforms to EN 60529)
Approvals/markings
Approvals/markings *)
CE, cURus
Additional tests
The devices have undergone the following additional tests:
Test
Explanation
Vibration
10 frequency sweeps in 3 axes
5 Hz < f < 60 Hz displacement 0.35 mm, constant amplitude
60.1 Hz < f < 500 Hz acceleration 5 g, constant amplitude
Shocks
1000 shocks in each direction, in 3 axes
35 g, 11 ms
